[Histonet] Leica Prisma problem

MARR, Penelope (EAST SUSSEX HEALTHCARE NHS TRUST) penny.marr at nhs.net
Fri Sep 23 01:51:09 CDT 2022


I suggest an extended prime to ensure no air in the line then consider if more drops of xylene are needed on the slides.

Hope this helps.

Penny  :)

Penny Marr
Senior BMS

C/- Histology
Conquest Hospital
St Leonards-on-Sea TN37 7RD

penny.marr at nhs.net

0300 131 4500 ext 734914
0300 131 4914

-----Original Message-----
From: Paula <plucas at biopath.org>
Sent: 22 September 2022 17:46
To: 'Rathborne, Toni' <Toni.Rathborne at RWJBH.org>
Cc: histonet at lists.utsouthwestern.edu
Subject: Re: [Histonet] Leica Prisma problem

Hi Toni,

Yes, I'm sorry..it is a Sakura product. I have called tech support and they want to visit but hoping I can fix it myself quicker if it's a simple fix.

Thanks,
Paula



-----Original Message-----
From: Rathborne, Toni [mailto:Toni.Rathborne at RWJBH.org]
Sent: Thursday, September 22, 2022 9:36 AM
To: 'Paula'
Subject: RE: [Histonet] Leica Prisma problem


Paula,

Isn't the Prisma a Sakura product? Have you called the vendor to ask for technical support over the phone? They are usually good about doing this, even if you don't have a service contract. You should also mention the lot number of the tape that you're using so they can document and track the problem if it has happened to other users.

Toni Rathborne
Pathology Supervisor
Robert Wood Johnson University Hospital Somerset



------------------------------------------------------------------------------------

NOTICE: This e-mail and its attachments, if any, may contain legally privileged and/or confidential information protected by law. It is intended only for use by the named addressee(s). If you are not the intended recipient of this e-mail, any dissemination, distribution or copying of this e-mail and its attachments, if any, is strictly prohibited. If you have received this transmission in error, please immediately notify the sender by telephone or by reply e-mail, and permanently delete this e-mail and the attachments, if any, and destroy any printouts.

-----Original Message-----
From: Paula via Histonet <histonet at lists.utsouthwestern.edu>
Sent: Thursday, September 22, 2022 11:56 AM
To: histonet at lists.utsouthwestern.edu
Subject: [Histonet] Leica Prisma problem


*** This is an External Email ***

Hello everyone,



We have the Leica Prisma film coverslipper and suddenly, there are bubbles underneath the film. I changed out the Xylene and inserted a new film roll but the problem persists.  Is there anything that you recommend to solve the problem?



Thank you in advance,

Paula Lucas

Lab Manager

Bio-Path Medical Group

_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.com/v3/__http://lists.utsouthwestern.edu/mailman/listinfo/histonet__;!!KU82p_qNqnQ!62MqADzSvCmn3BCClnb5s2_N3knTdw5FcHUsNGAHIOWQ9bhLFblOr-VTHzGGLyzIoxodLS-sW6xxh1lBS5Plhe6FKawFf-i9BDc$






************************************************************************************** ******************************

This message may contain confidential information. If you are not the intended recipient please:
i) inform the sender that you have received the message in error before deleting it; and
ii) do not disclose, copy or distribute information in this e-mail or take any action in relation to its content (to do so is strictly prohibited and may be unlawful).
Thank you for your co-operation.

NHSmail is the secure email, collaboration and directory service available for all NHS staff in England. NHSmail is approved for exchanging patient data and other sensitive information with NHSmail and other accredited email services.

For more information and to find out how you can switch visit Joining NHSmail – NHSmail Support<https://support.nhs.net/article-categories/joining-nhsmail/>



More information about the Histonet mailing list