[Histonet] Eosin Leaching
Vanessa Avalos
vavalos <@t> allergydermatology.com
Fri Jun 20 09:37:36 CDT 2014
We had the same problem. Increase your Alcohol/Reagents (following Eosin) times as well as using fresh ones. I use 80% for about 20 sec and (2)100% for a minute each and (3) Xylene Subs for 2 min each. I also use a fresh 80% and rotate the 100% each day. If using a Xylene Sub it could also be a possibility that your mounting medium is not compatible.
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Adrienne Anderson
Sent: Thursday, June 19, 2014 2:24 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Eosin Leaching
Hello all,
We're having a problem with our eosin bleeding out of the sections. I've read that this can be caused by not dehydrating adequately after eosin, but we're still having an issue when using fresh alcohols. Does anyone have any other ideas as to the cause of this problem?
Thanks in advance for your help!
Adrienne
Adrienne Anderson, BS, HTL(ASCP)
Histotechnologist
Phylogeny, Inc.
1476 Manning Pkwy, Powell, Ohio 43605
Phone: (614) 846-6161
Fax: (877) 591-1815
This message, including any attachments, is confidential and may be privileged or may contain health information protected by state and federal law.
Information and opinions expressed in this message and/or attachments are those of the author and are not necessarily those of the company.
If you are not the intended recipient, please notify the sender and delete this message from your system.
Any use of this information by individuals other than the intended recipient is strictly prohibited. _______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list