Arun,
It looks like you are overprocessing your tissues. Uterus and prostate are tissues that are often hard to cut and your schedule for processing may be adding to your difficulties.
Unless the pieces are very large and thick (and you can't do anything about that), cut down on the time, eliminate the heat and vacuum in your alcohols and xylenes but not the paraffins. You might also want to eliminate one of the 100% alcohols and one of the xylenes or try a xylene substitute like Clear-Rite 3, which is a more gentle clearing agent. If you eliminated one of these you could add another 70% or 80% alcohol before the 95% to help wash out the formalin. Right now you are really drying out your tissues and making them harder.
If your tissues aren't totally fixed you may need the time in Formalin but if they are well fixed you could cut these times also.
You probably don't need 4 hrs in paraffin at 63 degrees either. I'd cut the time down and the temp. to at most 60 degrees. I usually go 3-4 degrees above the melting point of the paraffin.

Andi Grantham

On Jan 25, 2012, at 6:25 AM, Arun Jyothi S.P wrote:

> Dear all,
> 
> We are having a hard time in cutting blocks. All our blocks especially
> uterus, prostate chips etc became very hard after processing, we cant even
> trim the blocks its very hard.
> we use SLEE MTM automated tissue processor
> and our schedule is
> 1, Formalin                            - 1hr          - ambient  - 40 degree
> 2, Formalin                            - 2 hr          - vacuum - 40 degree
> 3, 70% ethanol                       - 1hr          - vacuum  - Room temp
> 4, 95% ethanol                       - 1hr           - vacuum  - Room temp
> 5 and 6- 100% ethanol            - 1hr each   - vacuum - room temp
> 7, 100% ethanol                     - 1.30 hr      - vacuum  -  room temp
> 8, 9 and 10 - Xylene                - 1 hr each  - vacuum  - 30, 35, 40
> degree respectively
> 11,12,13 and 14 - paraffin wax  - 1 hr each  - vacuum -  63 degree for all.
> 
> We use Merck paraffin wax of 56 degree M.P
> 
> Is there any problem with tissue processing schedule or something else is
> wrong?
> 
> We are in a big problem really expecting some valuable suggestion and
> advices.
> 
> with regards
> 
> 
> 
> ARUN JYOTHI S.P.
> Histotechnologist
> United Laboratories Co. <http://www.unitedlabs.com.kw/>
> Kuwait
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 



More information about the Histonet mailing list