[Histonet] IHC pos. & neg. control question
Shea's
jshea121 <@t> roadrunner.com
Thu May 19 20:11:01 CDT 2011
Curt is right, according to CAP ...
ANP.22570 QC - Antibodies Phase II
Appropriate negative controls are used.
NOTE: Negative controls must assess the presence of nonspecific staining in patient tissue as well
as the specificity of each antibody. Results of controls must be documented, either in internal
laboratory records, or in the patient report. A statement in the report such as, "All controls show
appropriate reactivity" is sufficient.
A negative reagent control is used to assess nonspecific or aberrant staining in patient tissue related
to the antigen retrieval conditions and/or detection system used. A separate section of patient tissue
is processed using the same reagent and epitope retrieval protocol as the patient test slide, except
that the primary antibody is omitted, and replaced by any one of the following:
? An unrelated antibody of the same isotype as the primary antibody (for monoclonal
primary antibodies)
? An unrelated antibody from the same animal species as the primary antibody (for
polyclonal primary antibodies)
? The negative control reagent included in the staining kit
? The diluent/buffer solution in which the primary antibody is diluted
In general, a separate negative reagent control should be run for each block of patient tissue being
immunostained; however, for cases in which there is simultaneous staining of multiple blocks from
the same specimen with the same antibody (e.g. cytokeratin staining of multiple axillary sentinel
lymph nodes), performing a single negative control on one of the blocks may be sufficient provided
that all such blocks are fixed and processed identically. This exception does not apply to stains on
different types of tissues or those using different antigen retrieval protocols or antibody detection
systems. The laboratory director must determine which cases will have only one negative reagent
control, and this must be specified in the department's procedure manual.
The negative reagent control would ideally control for each reagent protocol and antibody retrieval
condition; however, large antibody panels often employ multiple antigen retrieval procedures. In
such cases, a reasonable minimum control would be to perform the negative reagent control using
the most aggressive retrieval procedure in the particular antibody panel. Aggressiveness of antigen
retrieval (in decreasing order) is as follows: pressure cooker; enzyme digestion, boiling; microwave;
steamer; water bath. High pH retrieval should be considered more aggressive than comparable
retrieval in citrate buffer at pH 6.0.
It is also important to assess the specificity of each antibody by a negative tissue control, which
must show no staining of tissues known to lack the antigen.The negative tissue control is processed
using the same fixation, epitope retrieval and immunostaining protocols as the patient tissue.
Unexpected positive staining of such tissues indicates that the test has lost specificity, perhaps
because of improper antibody concentration or excessive antigen retrieval. Intrinsic properties of
the test tissue may also be the cause of "non-specific" staining. For example, tissues with high
endogenous biotin activity such as liver or renal tubules may simulate positive staining when using
a detection method based on biotin labeling.
A negative tissue control must be processed for each antibody in a given run. Any of the following
can serve as a negative tissue control:
1. Multitissue blocks.These can provide simultaneous positive and negative tissue controls,
and are considered "best practice" (see below).
2. The positive control slide or patient test slides, if these slides contain tissue elements
that should not react with the antibody.
3. A separate negative tissue control slide.
The type of negative tissue control used (i.e. separate sections, internal controls or multitissue
blocks) should be specified in the laboratory manual (refer to ANP.22250).
Multitissue blocks may be considered best practice and can have a major role in maintaining quality.
When used as a combined positive and negative tissue control as mentioned above, they can serve
as a permanent record documenting the sensitivity and specificity of every stain, particularly when
mounted on the same slide as the patient tissue. When the components are chosen appropriately,
multitissue blocks may be used for many different primary antibodies, decreasing the number of
different control blocks needed by the laboratory. Multitissue blocks are also ideal for determining
optimal titers of primary antibodies since they allow simultaneous evaluation of many different pieces
of tissue. Finally, they are a useful and efficient means to screen new antibodies for sensitivity and
specificity or new lots of antibody for consistency, which should be done before putting any antibody
into diagnostic use.
Evidence of Compliance:
? Written procedure for the use of negative reagent and tissue controls for IHC AND
? Patient reports or worksheet with control results
More information about the Histonet
mailing list