[Histonet] Peloris Users
Josie Britton
JCBRITTON <@t> Cheshire-Med.COM
Thu Jul 8 11:03:08 CDT 2010
We have helped that problem by being sure there is adequate graded
alcohols on the Peloris. There must be between 60-70%, 80-90%, and 95%
before you have the 100% alcohols. If the concentration's are too high
there is inadequate infiltration of the tissue's. We usually only use
the 2 hour and the 8 hour processing factory default. Unless, we have a
stat. We have also noticed if you wait until Prompted to change the
lowest alcohol at 51% we see degradation in our staining as well. We
are not letting it get below 58-60% before we change an alcohol out.
I hope this helps,
Josie Britton HT
Cheshire Medical Center
580 Court Street
Keene, NH 03431
603-354-5454 ext.2454
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of
Sharon.Davis-Devine
Sent: Thursday, July 08, 2010 9:51 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Peloris Users
To all you Peloris users, here is my delimma. When we follow the
machines prompt to change the solutions because of purity and we then
process small biopsies on the machine, it overprocesses the specimens.
So the staff have been overrriding the prompt to change the solutions
causing specimens to be processed in impure solutions causing a host of
other problems. Have any of you out there with Peloris experience had
this problem? Is it a machine problem that should be fixed or human
error due to overriding the prompts? I am new to this position and have
very little experience with this issue so any help or suggestions will
be greatly appreiciated. Thank you.
Sharon Davis-Devine, CT (ASCP)
Cytology-Histology Supervisor
Carle Foundation Hospital
Laboratory and Pathology Services
611 West Park Street
Urbana, Illinois 61801
217-383-3572
sharon.davis-devine <@t> carle.com
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
CONFIDENTIALITY NOTICE: This electronic message, including any attachments, is for the sole use of the intended recipients and may contain confidential and privileged information. Any unauthorized review, use, disclosure or distribution is prohibited. If you are not the intended recipient, please contact the sender by electronic mail and destroy all copies of the original message.
More information about the Histonet
mailing list