Good Day, Tim:

I'll second what Terri said. My first concern with embedding a contact lens would be that the embedding media and/or transitional solvents would attack the lens and/and dislodge the bacteria. Unless you have a few lenses to do a method development pilot study with, I'd shy away from a full embedding. If you had to go the TEM route, maybe a plastic like LR White, as the dehydration doesn't have to be as aggressive.

The SEM is probably the best tool for this sort of investigation; we've had luck with identifying bacterial strains on morphology alone with our SEM, or at least it got us into the ballpark enough for the investigation.
Children's Colorado has an SEM in their facility.    

Cheers,
-Chris

Christopher Hayden
Electron Microscopy / Digital Imaging
LS - Specialty Laboratories
T +1 862 778 7993
christopher.hayden at novartis.com
Novartis Institutes for Biomedical Research



Date: Fri, 9 Mar 2018 19:30:27 +0000
From: "Terri  Braud" <tbraud at holyredeemer.com>
To: "'histonet at lists.utsouthwestern.edu'"
	<histonet at lists.utsouthwestern.edu>
Subject: Re: [Histonet] Identification of bacteria on an acrylic lens
	prosthesis
Message-ID:
	<48E053DDF6CE074DB6A7414BA05403F8B2325A46 at HRHEX02-HOS.holyredeemer.local>
	
Content-Type: text/plain;	charset="us-ascii"

I'm certified in EM. My suggestion would be to use SEM as opposed to TEM. The scanning electron microscope (SEM) can also be useful to reveal morphological features of isolated organisms as well as for diagnosis.  There are numerous articles citing this technique. I believe the UC Berkley has an SEM in their lab.    I hope this helps.  Terri

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Laboratory
Holy Redeemer Hospital
1648 Huntingdon Pike
Meadowbrook, PA 19046
ph: 215-938-3689
fax: 215-938-3874
Care, Comfort, and Heal

Today's Topics:
   1. Prosthetic lens processing, cutting? (Morken, Timothy)
Message: 1
Date: Thu, 8 Mar 2018 19:30:35 +0000
From: "Morken, Timothy" <Timothy.Morken at ucsf.edu>
To: Histonet <histonet at lists.utsouthwestern.edu>
Subject: [Histonet] Prosthetic lens processing, cutting?


Hi all,
Does anyone have any experience with embedding an acrylic prosthetic lens for EM, or any other way ? We have a case where they wanted EM on a lens to identify a  specific bacteria. We have never done it and it would take some work to figure it out. Also, I told them that unless there is some morphologic marker for the bacteria, EM would not help much identifying it.
Any ideas?
Any labs that handle this kind of thing? They are willing to send it out.
Tim Morken
Pathology Site Manager, Parnassus
Supervisor, Electron Microscopy/Neuromuscular Special Studies Department of Pathology UC San Francisco Medical Center



Christopher Hayden
T +1 862 778 7993
christopher.hayden at novartis.com


-----Original Message-----
From: histonet-request at lists.utsouthwestern.edu [mailto:histonet-request at lists.utsouthwestern.edu] 
Sent: Saturday, March 10, 2018 1:00 PM
To: histonet at lists.utsouthwestern.edu
Subject: Histonet Digest, Vol 172, Issue 7

Send Histonet mailing list submissions to
	histonet at lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
	https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=
or, via email, send a message with subject or body 'help' to
	histonet-request at lists.utsouthwestern.edu

You can reach the person managing the list at
	histonet-owner at lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific than "Re: Contents of Histonet digest..."


Today's Topics:

   1. Re: Histonet Digest, Vol 172,	Issue 6 "Slide labels or slide
      Printer" (Brent Adams)
   2. Re: Identification of bacteria on an acrylic lens	prosthesis
      (Terri  Braud)
   3. SAVE THE DATE!   HISTOPALOOZA  2018  APRIL 20-APRIL 22
      (Zimmerman, Billie)
   4. What is the smallest diameter (easily available and	cheap)
      syringe? (Jorge A. Santiago-Blay)


----------------------------------------------------------------------

Message: 1
Date: Fri, 9 Mar 2018 18:48:14 +0000
From: Brent Adams <badams at acadianagastro.com>
To: "histonet at lists.utsouthwestern.edu"
	<histonet at lists.utsouthwestern.edu>
Subject: Re: [Histonet] Histonet Digest, Vol 172,	Issue 6 "Slide
	labels or slide Printer"
Message-ID:
	<BLUPR03MB1380D38733C510F5A0D19EF2B1DE0 at BLUPR03MB1380.namprd03.prod.outlook.com>
	
Content-Type: text/plain; charset="Windows-1252"

I also have used Sakura Smart Write (Primera printers repackaged under the Sakura name)  slide and cassette printers since they came out. I had some issues with the slide printers in the beginning but they have worked out all the bugs and greatly improved the machine. I do very little maintenance other than clean rollers on the slide printer and never had any issues with the cassette printer. I recommend them for Small Labs for sure. They have multi cassette holders for larger labs.


Hope that helps

Brent Adams ? BS, LPN, HT


https://urldefense.proofpoint.com/v2/url?u=http-3A__www.acadianagastro.com&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=Dzzc938ChQJl3YEQb3wQdhPZLoBnOMy2id1HRrn_bLc&e=

Acadiana Gastroenterology Associates, LLC
439 Heymann Blvd
Lafayette, LA  70503

tel:  (337) 269-1126
fax:  (337) 269-1476


________________________________
From: histonet-request at lists.utsouthwestern.edu <histonet-request at lists.utsouthwestern.edu>
Sent: Friday, March 9, 2018 12:00 PM
To: histonet at lists.utsouthwestern.edu
Subject: Histonet Digest, Vol 172, Issue 6

Send Histonet mailing list submissions to
        histonet at lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
        https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=
or, via email, send a message with subject or body 'help' to
        histonet-request at lists.utsouthwestern.edu

You can reach the person managing the list at
        histonet-owner at lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific than "Re: Contents of Histonet digest..."


Today's Topics:

   1. Prosthetic lens processing, cutting? (Morken, Timothy)
   2. Re: Slide labels or slide labeler (Blazek, Linda)


----------------------------------------------------------------------

Message: 1
Date: Thu, 8 Mar 2018 19:30:35 +0000
From: "Morken, Timothy" <Timothy.Morken at ucsf.edu>
To: Histonet <histonet at lists.utsouthwestern.edu>
Subject: [Histonet] Prosthetic lens processing, cutting?
Message-ID:
        <BY1PR0501MB1352CDD8577A5DF78F749141E7DF0 at BY1PR0501MB1352.namprd05.prod.outlook.com>

Content-Type: text/plain; charset="us-ascii"

Hi all,


Does anyone have any experience with embedding an acrylic prosthetic lens for EM, or any other way ? We have a case where they wanted EM on a lens to identify a  specific bacteria. We have never done it and it would take some work to figure it out. Also, I told them that unless there is some morphologic marker for the bacteria, EM would not help much identifying it.

Any ideas?

Any labs that handle this kind of thing? They are willing to send it out.

Tim Morken
Pathology Site Manager, Parnassus
Supervisor, Electron Microscopy/Neuromuscular Special Studies Department of Pathology UC San Francisco Medical Center



------------------------------

Message: 2
Date: Thu, 8 Mar 2018 15:32:57 -0500
From: "Blazek, Linda" <lblazek at digestivespecialists.com>
To: Victor <victor_tobias at comcast.net>, "Vickroy, James"
        <jvickroy at SpringfieldClinic.com>
Cc: "histonet at lists.utsouthwestern.edu"
        <histonet at lists.utsouthwestern.edu>
Subject: Re: [Histonet] Slide labels or slide labeler
Message-ID:
        <5A2BD13465E061429D6455C8D6B40E392A826E8032 at IBMB7Exchange.digestivespecialists.com>

Content-Type: text/plain; charset="utf-8"

I've used the Primera slide printer since it came out.  I like it a lot.  It needs it's maintenance kept up like anything else but other than cleaning the rollers there is very little else that needs cleaning.

Linda Blazek HT (ASCP)
Pathology Lab Manager
GI Pathology of Dayton
Digestive Specialists, Inc
Phone: (937) 396-2623
Email: lblazek at digestivespecialists.com


-----Original Message-----
From: Victor via Histonet [mailto:histonet at lists.utsouthwestern.edu]
Sent: Wednesday, March 07, 2018 5:59 PM
To: Vickroy, James; histonet at lists.utsouthwestern.edu
Subject: Re: [Histonet] Slide labels or slide labeler

James,

I totally agree with your assessment of the Slidemate printers. Have not had any experience with the newer model.
Applying slide labels shouldn?t be an issue when dealing with one block and slide(s) at a time.  Some of this could depend upon your software available to you. Our system generated just the labels needed for the block you were cutting, on demand. You should be able to get a nice Zebra printer for under $700.

Another of our labs used the new Sakura/Premera printer and really like it.

Victor

Sent from Mail for Windows 10

From: Vickroy, James via Histonet
Sent: Wednesday, March 7, 2018 2:12 PM
To: histonet at lists.utsouthwestern.edu
Subject: [Histonet] Slide labels or slide labeler

We have used the older version Slidemate slide labelers from Thermofisher for the last three years.   I think putting the patient information directly on the slide is preferred however the wear and tear on slide printers means higher maintenance costs and more down time.  Thermofisher has a newer version of the Slidemate now that has been designed so that there is less wear and tear on the machines.   Currently when one of the old labelers breaks down we have to send the instrument out and the cost is generally 1000 - 1500 for repair each time.    It also seems that the older machines have to be sent in at least every other year.  Our older machines have been sent in almost yearly.

We are considering switching to a slide label making system by General Data.   One of the downsides is that we have to apply the label to each slide so therefore we introduce another step where the slides could be mislabeled.   The upside is that equipment costs are much less.   A new thermal label printer runs around 700.00 where as the instrument that imprints the slide label on the slide runs between 12000 - 15000 per unit.  We need three units.
Thermo also offers a lease system.    What I need to know is what has been the experience with the new Slidemate AS printers?   What kind of maintenance issues have you encountered? Finally please let me know your latest experiences with both slide label printers versus slide labeling printers?

Thanks for your input.

Jim


Jim Vickroy
Histology Manager
Springfield Clinic, Main Campus, East Building
1025 South 6th Street
Springfield, Illinois  62703
Office:  217-528-7541, Ext. 15121
Email:  jvickroy at SpringfieldClinic.com<mailto:jvickroy at SpringfieldClinic.com>



This electronic message contains information from Springfield Clinic, LLP that may be confidential, privileged, and/or sensitive. This information is intended for the use of the individual(s) or entity(ies) named above. If you are not the intended recipient, be aware that disclosure, copying, distribution, or action taken on the contents of this information is strictly prohibited. If you have received this electronic message in error, please notify the sender immediately, by electronic mail, so that arrangements may be made for the retrieval of this electronic message. Thank you.
_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=

_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=

Histonet Info Page - https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=1sE1SSoyO-runtuK_WoVchPnoSApq7C-coCsWs8y0gg&e= Mailing Lists<https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=1sE1SSoyO-runtuK_WoVchPnoSApq7C-coCsWs8y0gg&e=
Histonet -- For the exchange of information pertaining to histotechnology and related fields About Histonet




------------------------------

Subject: Digest Footer

_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=

Histonet Info Page - https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=1sE1SSoyO-runtuK_WoVchPnoSApq7C-coCsWs8y0gg&e= Mailing Lists<https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=1sE1SSoyO-runtuK_WoVchPnoSApq7C-coCsWs8y0gg&e=
Histonet -- For the exchange of information pertaining to histotechnology and related fields About Histonet




------------------------------

End of Histonet Digest, Vol 172, Issue 6
****************************************
PRIVILEGED AND CONFIDENTIAL: This document and the information contained herein are confidential and protected from disclosure pursuant to Federal Law. This message is for the designated recipient only and may contain confidential, privileged, proprietary, or otherwise private information. If you have received this email in error, please notify the sender immediately and delete the original with any attachments. Any other use of the email is strictly prohibited.


------------------------------

Message: 2
Date: Fri, 9 Mar 2018 19:30:27 +0000
From: "Terri  Braud" <tbraud at holyredeemer.com>
To: "'histonet at lists.utsouthwestern.edu'"
	<histonet at lists.utsouthwestern.edu>
Subject: Re: [Histonet] Identification of bacteria on an acrylic lens
	prosthesis
Message-ID:
	<48E053DDF6CE074DB6A7414BA05403F8B2325A46 at HRHEX02-HOS.holyredeemer.local>
	
Content-Type: text/plain;	charset="us-ascii"

I'm certified in EM. My suggestion would be to use SEM as opposed to TEM. The scanning electron microscope (SEM) can also be useful to reveal morphological features of isolated organisms as well as for diagnosis.  There are numerous articles citing this technique. I believe the UC Berkley has an SEM in their lab.    I hope this helps.  Terri

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Laboratory
Holy Redeemer Hospital
1648 Huntingdon Pike
Meadowbrook, PA 19046
ph: 215-938-3689
fax: 215-938-3874
Care, Comfort, and Heal

Today's Topics:
   1. Prosthetic lens processing, cutting? (Morken, Timothy)
Message: 1
Date: Thu, 8 Mar 2018 19:30:35 +0000
From: "Morken, Timothy" <Timothy.Morken at ucsf.edu>
To: Histonet <histonet at lists.utsouthwestern.edu>
Subject: [Histonet] Prosthetic lens processing, cutting?


Hi all,
Does anyone have any experience with embedding an acrylic prosthetic lens for EM, or any other way ? We have a case where they wanted EM on a lens to identify a  specific bacteria. We have never done it and it would take some work to figure it out. Also, I told them that unless there is some morphologic marker for the bacteria, EM would not help much identifying it.
Any ideas?
Any labs that handle this kind of thing? They are willing to send it out.
Tim Morken
Pathology Site Manager, Parnassus
Supervisor, Electron Microscopy/Neuromuscular Special Studies Department of Pathology UC San Francisco Medical Center


***********************
---------------------------------------------------------------------------------



CONFIDENTIALITY NOTICE:

This E-Mail is intended only for the use of the individual or entity to which it was sent. It may contain information that is privileged and/or confidential, and the use or disclosure of such information may also be restricted under applicable federal and state law. If you received this communication in error, please do not distribute any part of it or retain any copies, and delete the original E-Mail.
Please notify the sender of any error by E-Mail.

Thank you for your cooperation.




------------------------------

Message: 3
Date: Fri, 9 Mar 2018 19:43:06 +0000
From: "Zimmerman, Billie" <BZIMMERM at augusta.edu>
To: "histonet at lists.utsouthwestern.edu"
	<histonet at lists.utsouthwestern.edu>
Subject: [Histonet] SAVE THE DATE!   HISTOPALOOZA  2018  APRIL
	20-APRIL 22
Message-ID:
	<BN6PR03MB25946D3A4DC67341A904F832CBDE0 at BN6PR03MB2594.namprd03.prod.outlook.com>
	
Content-Type: text/plain; charset="us-ascii"

TGIF everyone
The Georgia Society for Histotechnology meeting will be at the Callaway Gardens Lodge, Pine Mountain, GA More details to follow, but for now save the date!

Billie Zimmerman


------------------------------

Message: 4
Date: Sat, 10 Mar 2018 10:43:45 -0500
From: "Jorge A. Santiago-Blay" <blayjorge at gmail.com>
To: histonet at lists.utsouthwestern.edu
Subject: [Histonet] What is the smallest diameter (easily available
	and	cheap) syringe?
Message-ID:
	<CAGBdDuArmduaFBYTDLe-35ErFLYzSycud7j1WVXuGzPoct6bcw at mail.gmail.com>
Content-Type: text/plain; charset="UTF-8"

Hi:

I am preparing a device for blowing air into a tiny, subcylindrical structure (diameter << 1 mm) in some (non-human) organisms. Heat extruded Pasteur pipettes are good but too fragile for my taste.

What  is the smallest diameter (easily available and cheap) syringe?

If you have any constructive suggestions, please email me directly.

Cheers,

Jorge

Jorge A. Santiago-Blay, PhD
blaypublishers.com

1. Positive experiences for authors of papers published in *LEB* https://urldefense.proofpoint.com/v2/url?u=http-3A__blaypublishers.com_testimonials_&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=_3zzySWXoTSNwwhKP1OATmU-uVIe0lp9_j3cOdLnARQ&e=

2. Free examples of papers published in *LEB*:
https://urldefense.proofpoint.com/v2/url?u=http-3A__blaypublishers.com_category_previous-2Dissues_&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=C2Dlouw0YUh511iEBtc1kkvBoVND1_35YhsDcpqK2Xw&e=.

3. *Guidelines for Authors* and page charges of *LEB*:
https://urldefense.proofpoint.com/v2/url?u=http-3A__blaypublishers.com_archives_&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=uehOhIvCm7C4YyqH_tqCq4q7tR6HZudXwWthWI1cmjE&e= *.*

4. Want to subscribe to *LEB*? https://urldefense.proofpoint.com/v2/url?u=http-3A__blaypublishers.com_subscriptions_&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=tAUxfakrhWzEwyxfhO1OibaV9k5B55v1EHneJzSghY8&e=


https://urldefense.proofpoint.com/v2/url?u=http-3A__blayjorge.wordpress.com_&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=2oRa8t3wEZnHmETD4JeMstw_Oy10qujP3i7Jz64RYwQ&e=
https://urldefense.proofpoint.com/v2/url?u=http-3A__paleobiology.si.edu_staff_individuals_santiagoblay.cfm&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=9Y3WUcIUlKAiofJQuytYrCrlaG5-MbCVFDos6UhK-ww&e=


------------------------------

Subject: Digest Footer

_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet&d=DwICAg&c=ZbgFmJjg4pdtrnL2HUJUDw&r=oHNFb0vNgGhlfuzljZEWz1In4NMRNTEnF8z_Nz6USno&m=VFtPXj_ahjTodptDNNpwf48ccKM7L6xQMCWbtd7jFXU&s=MatAxjSKXHZSS171h5q_lpGJSc4DpJw9cmQV2a37oU8&e=

------------------------------

End of Histonet Digest, Vol 172, Issue 7
****************************************



More information about the Histonet mailing list