[Histonet] temperature charts

Bryan, Karen KBRYAN1 at PARTNERS.ORG
Wed Feb 3 13:29:22 CST 2016


Hi!

We have a temperature chart form that we created and use across all instruments and areas. It is one chart for the whole year and has an area for review date and signature at the end of each month.
We have daily thermometers. These may be regular thermometers or built in to the instrument. Twice a year we check those daily thermometers against our "gold standard" calibration thermometer. The one we use is a Fisher "Ertco" Standard. Min/Max daily thermometers are also cross checked even if they come with their own calibration report. If the daily is off by +/- 1 deg. C then we replace the daily and monitor it to be sure it is not the instrument that is faulty.
We collect all temperature QC Charts and Calibration reports at the end of each year for JCAHO inspection.

Hope that helps!

Thank you,

Karen

Karen P. Bryan, HT (ASCP)cm
Sr. Neurohistology Specialist
Department of Pathology/Amory 3
Brigham & Women's Hospital
75 Francis Street/Boston
617-732-7534
(Internal extension 2-7534)

Karen


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.