[Histonet] H&E Stainer Question
WILLIAM DESALVO
wdesalvo.cac at outlook.com
Tue May 12 19:10:46 CDT 2015
Look at all the automated H&E stain instruments on the market. I suggest that you consider those that offer the best benefit to your workload and workflow. Instruments that utilize the concept of co-location (related tasks linked together; oven fro drying, flexible stain configuration and Coverslipping) will assist you in developing a LEAN workflow. Consider how many times you need to touch the slides to complete all the tasks and how much walk away time you gain.
I do not suggest by-passing the validated staining instrument oven. Placing slides in another oven creates variation and often results in short and extended drying times. All automated H&E stain instruments should be used according to manufacturer recommendation. My experience is that when shortcuts are used, quality suffers. The automated H&E stain instruments have great through put and you should adjust workflow to maximize the designated batch size and stain time.
Film coverslip lasts longer than 7 years. Sakura film is the best and has testing to exceed 10 years. I have used it for 13 years and never had the film peel. If you use the knock off film products, they only have short term accelerated stability testing and probably do not have any version of their many changes to the emulsion that have real time stability testing to exceed 10 years. There are reasons to use glass and reasons to use film. Both are great products, when you use them correctly and purchase quality products. Glass automated cover slip options on instruments do have more required maintenance than film.
CAP has made no statement about film cover slipping. In fact, the Hologic (was Cytech) Cytology Thinprep system is FDA approved with film as the cover slip and the stained and film cover slipped Thinprep slide is digitally scanned for analysis. To be CAP compliant, you must keep blocks and slides for 10 years. With the advancement of cancer hospital protocols and molecular testing, many institutions are considering retaining blocks and slides beyond 10 years. Keeping blocks and slides longer than regulation requires introduces a large risk factor for the retaining institution and pathologists.
Always be forward thinking when considering the purchase of a new essential instrument. Will it bar code read, can it be interfaced to LIS or tracking system, what analytics can be extract and will the instrument help or hinder a LEAN workflow? There are many choices that will meet your basic needs, but which one meets your essential needs?
William DeSalvo, BS HTL(ASCP)
> From: simmca at UPMC.EDU
> To: Royl1 at LabCorp.com; patpxs at gmail.com; histonet at lists.utsouthwestern.edu
> Date: Tue, 12 May 2015 12:32:43 +0000
> Subject: Re: [Histonet] H&E Stainer Question
>
> To be fair, a "batch" of slides for Leica is actually 270- slides, It can run 9 racks at a time, but, the 9th rack from start to coverslip is 3 hours+
> You can always skip the on-board oven and place your slides in a slide dryer (most labs have them) and then every 3 minutes you can load a rack (1st xylene step 3 minutes)
> Then it goes much faster.
> As for tape..ugh..it is only guaranteed to last 7 years, after that they pull off the slide and take the tissue with it.
> CAP is starting to frown on this as you need to keep the initial H&E slides for up to 10+ years.
>
> Chris Simmons B.S., A.S., HTL(ASCP)
> Supervisor, UPP Dermatopathology
> 412.864.3880 office
> 412.612.0881 cell
>
>
> -----Original Message-----
> From: Roy, Lisa [mailto:Royl1 at LabCorp.com]
> Sent: Tuesday, May 12, 2015 8:19 AM
> To: Paula Sicurello; HistoNet
> Subject: Re: [Histonet] H&E Stainer Question
>
> Paula
> Here are my two cents....
>
> I currently use a Leica Autostainer XL with attached glass coverslipper. It is consistent in its staining and easy to use. The downfall is if you are a large volume lab or just have large volume days, each staining rack holds 30 slides and only one rack can be stained in each batch. The stainer also only has one on board oven, so the throughput of this machine is fairly low. It is only staining 30 slides at a time, with one holding station for the next set. It will run multiple batches concurrently, but gets to a point where it is all backed up. We sometimes have 2 racks staining, one in the oven, one in the loading dock, and some sitting on top of the stainer until it can go on.
>
> On the flip side, the Sakura Prisma is a workhorse. It is very similar to the Leica in the sense that it is linear and very consistent in staining. It has two on board ovens and each basket can hold 20 slides. The difference is that the Sakura can stain 3 racks (60 slides) per batch, with two batches in the oven at the same time. That gives you 120 slide throughput for each batch. This stainer also has an attached coverslipper (Sakura Film), but it is film coverslips. I know, I know.....no one likes the film coverslips. One advantage to the film, is that the slides are dry almost immediately and can be filed away the same day. No waiting for 3-4 days for the glass ones to fully cure. I can say that the last lab I worked in had the film coverslips and after 10 years, the slides were still in pristine condition. There are many pathologists that do not like to read film covered slides, but once ours got used to it, they had no problems. Some say the film yellows or comes off with the tissue still attached. I can say that I never seen this in my past position. It is very dependent on Xylene only during coverslipping. You cannot use a xylene substitute in the coverslip portion and expect to get good results.
>
> Hope it helps....good luck.
> Lisa
>
> -----Original Message-----
> From: Paula Sicurello [mailto:patpxs at gmail.com]
> Sent: Monday, May 11, 2015 5:04 PM
> To: HistoNet
> Subject: [Histonet] H&E Stainer Question
>
> Me again...
>
> UCSD is in the market for a new H&E stainer for our new hospital opening next year.
>
> We need a workhorse, not a prima dona, something with a coverslipper built in would be nice.
>
> What do you use?
>
> Suggestions gratefully accepted-even from you two Keith and Matt ;)
>
> Opinions about the good, the bad, and the ugly (as long as it works really
> well) will be helpful.
>
> Thanks oodles!
>
> Paula :-)
> _______________________________________________
> Histonet mailing list
> Histonet at lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> -This e-mail and any attachments may contain CONFIDENTIAL information, including PROTECTED HEALTH INFORMATION. If you are not the intended recipient, any use or disclosure of this information is STRICTLY PROHIBITED; you are requested to delete this e-mail and any attachments, notify the sender immediately, and notify the LabCorp Privacy Officer at privacyofficer at labcorp.com or call (877) 23-HIPAA / (877) 234-4722.
>
>
> _______________________________________________
> Histonet mailing list
> Histonet at lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
> _______________________________________________
> Histonet mailing list
> Histonet at lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list