Hi Wes,
We have 3 bond RX units and use them for animal tissue staining and we try all kinds of reagents (ab, chromogens, counterstains, acid washes) . As far as the HEIR and Enzyme go Leica has their own HEIR (citrate & EDTA) and an Enzyme (PTK). You can run a Dako PTK which I do at room temp or at 37C but as far as the HEIR you are limited to their Citrate Ph= 6 and EDTA Ph=9, I think I'm right on the Ph's but not sure.
Anyway, we have found that all of our Dako protocols using PT modules or Pressure cooker transferred nicely to Leica and there recommended times for citrate and EDTA are pretty darn good. It is a learning curve and there are Leica ways of doing things but you can get to do almost anything you can do on the bench. We are working up a lot of double now and there are so many ways to do it.
Hope that helps.
Jamie
JAMIE ERICKSON
Scientist II
HTL (ASCP)CM,MS
Abbvie Bioreseach Center
Pharmacology
100 Research Dr.
Worcester,Ma 01605
OFFICE +1 508-688-3134
FAX +1 508-793-4895
EMAIL jamie.erickson <@t> abbvie.com
abbvie.com
This communication may contain information that is proprietary, confidential, or exempt from disclosure. If you are not the intended recipient, please note that any other dissemination, distribution, use or copying of this communication is strictly prohibited. Anyone who receives this message in error should notify the sender immediately by telephone or by return e-mail and delete it from his or her computer.
******************