[Histonet] RE: Low Temp problem

Tom McNemar TMcNemar <@t> lmhealth.org
Thu Oct 24 04:40:33 CDT 2013


We have had pads replaced as well but the problem continues.

Tom McNemar, HT(ASCP)
Histology Co-ordinator
Licking Memorial Health Systems
(740) 348-4163
(740) 348-4166
tmcnemar <@t> lmhealth.org
www.LMHealth.org


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Gary Gracie
Sent: Wednesday, October 23, 2013 9:29 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Low Temp problem

I am also running Ventana Benchmark Ultras. This low temp problem has started to occur regularly to individual heater pads (five in total over several weeks, across two machines). Ventana have promptly replaced the pad on each occasion.

Gary Gracie
Senior Technical Officer
IHC Laboratory
Anatomical Pathology
St Vincents Hospital
Sydney, Australia

Message: 1
Date: Tue, 22 Oct 2013 11:04:45 -0700
From: "HERRINGTON, SHEILA" <SHEILA.HERRINGTON <@t> interiorhealth.ca>
Subject: RE: [Histonet] Low Temp question
To: 'Tom McNemar' <TMcNemar <@t> lmhealth.org>, 'Shawn Leslie '
        <shawnleslie <@t> embarqmail.com>, "Histonet <@t> lists.utsouthwestern.edu"
        <Histonet <@t> lists.utsouthwestern.edu>
Message-ID:
        <9D5F3F245A5FE0458E46E9A71E3EE08C06CD37F350 <@t> DC1SERV352.interiorhealth.ca>

Content-Type: text/plain; charset="utf-8"

I also have been having low temp errors.  I use  Benchmark Ultras and they are not a batch stainer, so pads are heated independently.  I noticed the staining on the particular pad with the error was not quite to the usual standard so I have stopped using that pad.  Have contacted Ventana for a case number, but have not got an answer.  Is this a new problem for you as well?  This only started a few weeks ago for us.


Sheila Herrington
Technical Lead Histopathology and Immunohistochemistry
Kelowna General Hospital
2268 Pandosy Street, Kelowna, B.C. V1Y 1T2
250-862-4300 ext 7587 or 7510
Sheila.herrington <@t> interiorhealth.ca





-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Tom McNemar
Sent: Tuesday, October 22, 2013 8:28 AM
To: 'Shawn Leslie '; Histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Low Temp question

We see these continually as well.  We are told that it is a software glitch and not really a problem with the heat. I was told that they are working on it.

Tom McNemar, HT(ASCP)
Histology Co-ordinator
Licking Memorial Health Systems
(740) 348-4163
(740) 348-4166
tmcnemar <@t> lmhealth.org
www.LMHealth.org<http://www.LMHealth.org>


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Shawn Leslie
Sent: Tuesday, October 22, 2013 11:07 AM
To: Histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Low Temp question


Goodmorning all,
    I have a question concerning the Ventana Benchmark XT. We are constantly having low temperature errors during almost every run. The immunos look fine but we are still concerned. Has anyone else that uses the Benchmark experienced this also ? We have noticed that the silver coating on the heating pads is wearing off. Ventana indicated to us that it's nothing to be concerned about. Any hel p would be appreciated.

Shawn Leslie HT (ASCP)

**********************************************************************
This email and any files transmitted with it are confidential and
intended solely for the use of the individual or entity to whom they
are addressed. If you have received this email in error please notify
the system manager.

This footnote also confirms that this email message has been virus
scanned and although no viruses were detected by the system, St Vincents &
Mater Health Sydney accepts no liability for any consequential damage
resulting from email containing any computer viruses.

**********************************************************************
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

This e-mail, including attachments, is intended for the sole use of the individual and/or entity to whom it is addressed, and contains information from Licking Memorial Health Systems which is confidential or privileged. If you are not the intended recipient, nor authorized to receive for the intended recipient, be aware that any disclosure, copying, distribution or use of the contents of this e-mail and attachments is prohibited. If you have received this in error, please advise the sender by reply e-mail and delete the message immediately. You may also contact the LMH Process Improvement Center at 740-348-4641. E-mail transmissions cannot be guaranteed to be secure or error-free as information could be intercepted, corrupted, lost, destroyed, arrive late or incomplete, or contain viruses. The sender therefore does not accept liability for any errors or omissions in the contents of this message, which arise as a result of e-mail transmission. Thank you.



More information about the Histonet mailing list