I believe you may need to have the unit serviced.  It sounds like something is not tight enough, perhaps the stage or blade holder unit.  You said you secured everything which makes me think you have some issue with the cryostat itself.  If a sharp blade, tightened blade and specimen, and varying the speed doesn't work, then you may need professional assistance form a service technician, especially if it is an older unit that has been moved around.  Have you tried different temperatures?  I don't think that is the problem but it is worth a try. 

Good luck! Cryosectioning can be an art in itself. Those decent sections are likely thicker than you intend, by the way.  Classic "thick and thin".


Philip Manfre, B.A., HT (ASCP)
Associate Principal Scientist
Merck Research Laboratories
WP45-251
PO Box 4
West Point, PA 19486

215-652-9750
215-993-0383 (fax)
philip_manfre <@t> merck.com




-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Perow, Elliott S (PEROWES10)
Sent: Thursday, October 03, 2013 12:33 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Inconsistent Sections with Cryostat 

Hello All,

I have a Leica Cryocut 1800 with a Reichert-Jung 2020 microtome that has sat for 8 years and went through a move from one academic building to another.  I am trying to section brook trout brain, but am having difficulty getting consistent sections.
I will get a good section and then the next section will be a very very thin shaving of usually just the top portion of the OCT block.  I have tried adjusting the clearance angle, made sure the blade and specimen is secured, replaced the blade with a new one, have tried different sectioning speeds, different thicknesses of the section and still continually only get a decent section every other turn.  I was wondering if anyone has had any similar experiences or if maybe the machine just needs a maintenance check due to it being an older machine.  I wasn't sure if the advancement mechanism may be off or if it is another issue that doesn't seem apparent to me.  Any help would be greatly appreciated.

Thanks,

Elliott Perow
Juniata College
Biology POE


CONFIDENTIALITY NOTICE: The materials in this electronic mail transmission (including all attachments) are private and confidential and are the property of the sender. The information contained in the material is privileged and is intended only for the use of the named addressee(s). If you are not the intended addressee, be advised that any unauthorized disclosure, copying, distribution or the taking of any action in reliance on the contents of this material is strictly prohibited. If you have received this e-mail in error, please immediately notify the sender by replying to the e-mail, and then destroy it immediately. Thank you.

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
Notice:  This e-mail message, together with any attachments, contains
information of Merck & Co., Inc. (One Merck Drive, Whitehouse Station,
New Jersey, USA 08889), and/or its affiliates Direct contact information
for affiliates is available at 
http://www.merck.com/contact/contacts.html) that may be confidential,
proprietary copyrighted and/or legally privileged. It is intended solely
for the use of the individual or entity named on this message. If you are
not the intended recipient, and have received this message in error,
please notify us immediately by reply e-mail and then delete it from 
your system.




More information about the Histonet mailing list