Hi Janine, we use cassette printers but we still write the number of pieces on one side of the cassette and any special  instructions on the other side, such as embed on edge or embed on red ink, etc.

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of histonet-request <@t> lists.utsouthwestern.edu
Sent: Friday, February 08, 2013 1:01 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: Histonet Digest, Vol 111, Issue 13

Send Histonet mailing list submissions to
	histonet <@t> lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
	http://lists.utsouthwestern.edu/mailman/listinfo/histonet
or, via email, send a message with subject or body 'help' to
	histonet-request <@t> lists.utsouthwestern.edu

You can reach the person managing the list at
	histonet-owner <@t> lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific
than "Re: Contents of Histonet digest..."


Today's Topics:

   1. RE: Temperatures (Sullivan, Beatrice)
   2. RE: Temperatures (Tom McNemar)
   3. RE: Temperatures (Kuhnla, Melissa)
   4. RE: Cassette labeling (Podawiltz, Thomas)
   5. Re: RE: Temperatures (Paula Sicurello)
   6. PIN4 staining (Cathy Crumpton)


----------------------------------------------------------------------

Message: 1
Date: Fri, 8 Feb 2013 15:32:22 +0000
From: "Sullivan, Beatrice" <BSullivan <@t> virtua.org>
Subject: RE: [Histonet] Temperatures
To: Rene J Buesa <rjbuesa <@t> yahoo.com>,
	"histonet <@t> lists.utsouthwestern.edu"
	<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
	<6932520047F7EE46B512E9801344F1600467CC <@t> ExchangeMB-1.Virtua.org>
Content-Type: text/plain; charset="utf-8"

Because I have been told that according to JCAHO (who will be inspecting us) that if a temperature is required to ascertain consistency, quality and integrity, you will have to record temperature 7 days per week.

From: Rene J Buesa [mailto:rjbuesa <@t> yahoo.com]
Sent: Friday, February 08, 2013 10:17 AM
To: Sullivan, Beatrice; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] Temperatures

And my question is: why are you bothered by not knowing the temp. in a processor that only starts to process on Sundays?
If the processor is emptied on Mondays to embed→cut, take the temp. then.
Other that this you would have to place a sensor and a recording device or you would have to make somebody to go to the lab. For just write down the melted paraffin temp.? Too much expense and trouble for an non-existent problem.
René J.

From: "Sullivan, Beatrice" <BSullivan <@t> virtua.org<mailto:BSullivan <@t> virtua.org>>
To: "histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>" <histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>>
Sent: Friday, February 8, 2013 7:52 AM
Subject: [Histonet] Temperatures

I'm looking for a fix to our problem of no temperatures being taken on the weekends. We are closed and this is creating an issue. Our processors are not running until Sunday night but the paraffin in both the processors and embedding center are kept molten. Any help would be greatly appreciated

Beatrice L. Sullivan HT(ASCP)HTL
Corporate Histology Manager
Virtua, Voorhees
856-247-3144


This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org<mailto:ISSECURITY <@t> virtua.org>.

Thank you

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu<mailto:Histonet <@t> lists.utsouthwestern.edu>
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org. 

Thank you


------------------------------

Message: 2
Date: Fri, 8 Feb 2013 11:22:59 -0500
From: Tom McNemar <TMcNemar <@t> lmhealth.org>
Subject: RE: [Histonet] Temperatures
To: "'Sullivan, Beatrice'" <BSullivan <@t> virtua.org>, Rene J Buesa
	<rjbuesa <@t> yahoo.com>, "histonet <@t> lists.utsouthwestern.edu"
	<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
	<E9A90E28259D2F4E84308C5E8EA8F7B40138657302AF <@t> lmhs-exchange>
Content-Type: text/plain; charset="utf-8"

We are JCAHO and this has never come up but that doesn't mean that it won't.

I don’t know about your specific processor but ours (VIP5), will fail to start and alarm if the paraffin temp is out of range.  Our processor is tied in to the main lab temperature monitoring system so we get a call anytime there is an alarm.

Such a system is probably not feasible for a processor alone but does your main lab have anything in place?

Tom McNemar, HT(ASCP)
Histology Co-ordinator
Licking Memorial Health Systems
(740) 348-4163
(740) 348-4166
tmcnemar <@t> lmhealth.org
www.LMHealth.org


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Sullivan, Beatrice
Sent: Friday, February 08, 2013 10:32 AM
To: Rene J Buesa; histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Temperatures

Because I have been told that according to JCAHO (who will be inspecting us) that if a temperature is required to ascertain consistency, quality and integrity, you will have to record temperature 7 days per week.

From: Rene J Buesa [mailto:rjbuesa <@t> yahoo.com]
Sent: Friday, February 08, 2013 10:17 AM
To: Sullivan, Beatrice; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] Temperatures

And my question is: why are you bothered by not knowing the temp. in a processor that only starts to process on Sundays?
If the processor is emptied on Mondays to embed→cut, take the temp. then.
Other that this you would have to place a sensor and a recording device or you would have to make somebody to go to the lab. For just write down the melted paraffin temp.? Too much expense and trouble for an non-existent problem.
René J.

From: "Sullivan, Beatrice" <BSullivan <@t> virtua.org<mailto:BSullivan <@t> virtua.org>>
To: "histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>" <histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>>
Sent: Friday, February 8, 2013 7:52 AM
Subject: [Histonet] Temperatures

I'm looking for a fix to our problem of no temperatures being taken on the weekends. We are closed and this is creating an issue. Our processors are not running until Sunday night but the paraffin in both the processors and embedding center are kept molten. Any help would be greatly appreciated

Beatrice L. Sullivan HT(ASCP)HTL
Corporate Histology Manager
Virtua, Voorhees
856-247-3144


This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org<mailto:ISSECURITY <@t> virtua.org>.

Thank you

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu<mailto:Histonet <@t> lists.utsouthwestern.edu>
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org.

Thank you


This e-mail, including attachments, is intended for the sole use of the individual and/or entity to whom it is addressed, and contains information from Licking Memorial Health Systems which is confidential or privileged. If you are not the intended recipient, nor authorized to receive for the intended recipient, be aware that any disclosure, copying, distribution or use of the contents of this e-mail and attachments is prohibited. If you have received this in error, please advise the sender by reply e-mail and delete the message immediately. You may also contact the LMH Process Improvement Center at 740-348-4641. E-mail transmissions cannot be guaranteed to be secure or error-free as information could be intercepted, corrupted, lost, destroyed, arrive late or incomplete, or contain viruses. The sender therefore does not accept liability for any errors or omissions in the contents of this message, which arise as a result of e-mail transmission. Thank you.

------------------------------

Message: 3
Date: Fri, 8 Feb 2013 11:23:04 -0500
From: "Kuhnla, Melissa" <Melissa.Kuhnla <@t> chsli.org>
Subject: RE: [Histonet] Temperatures
To: "Sullivan, Beatrice" <BSullivan <@t> virtua.org>, Rene J Buesa
	<rjbuesa <@t> yahoo.com>, <histonet <@t> lists.utsouthwestern.edu>
Message-ID:
	<C76F12086768614F9C2618ED9966FEE10F3F8F9A <@t> MMDCNT0BXVS003.chsli.org>
Content-Type: text/plain; charset="gb2312"

Beatrice,
How about...your processor monitors the temp.  If it falls out of a certain range it will most likely throw an error.  This is not a recorded temp day 6 and 7 but is in a sense a monitoring system.  Our processors are set up with an alarm system that call employees if certain errors occur.  Do you have something similar?

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Sullivan, Beatrice
Sent: Friday, February 08, 2013 10:32 AM
To: Rene J Buesa; histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Temperatures

Because I have been told that according to JCAHO (who will be inspecting us) that if a temperature is required to ascertain consistency, quality and integrity, you will have to record temperature 7 days per week.

From: Rene J Buesa [mailto:rjbuesa <@t> yahoo.com]
Sent: Friday, February 08, 2013 10:17 AM
To: Sullivan, Beatrice; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] Temperatures

And my question is: why are you bothered by not knowing the temp. in a processor that only starts to process on Sundays?
If the processor is emptied on Mondays to embed¡úcut, take the temp. then.
Other that this you would have to place a sensor and a recording device or you would have to make somebody to go to the lab. For just write down the melted paraffin temp.? Too much expense and trouble for an non-existent problem.
Ren¨¦ J.

From: "Sullivan, Beatrice" <BSullivan <@t> virtua.org<mailto:BSullivan <@t> virtua.org>>
To: "histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>" <histonet <@t> lists.utsouthwestern.edu<mailto:histonet <@t> lists.utsouthwestern.edu>>
Sent: Friday, February 8, 2013 7:52 AM
Subject: [Histonet] Temperatures

I'm looking for a fix to our problem of no temperatures being taken on the weekends. We are closed and this is creating an issue. Our processors are not running until Sunday night but the paraffin in both the processors and embedding center are kept molten. Any help would be greatly appreciated

Beatrice L. Sullivan HT(ASCP)HTL
Corporate Histology Manager
Virtua, Voorhees
856-247-3144


This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org<mailto:ISSECURITY <@t> virtua.org>.

Thank you

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu<mailto:Histonet <@t> lists.utsouthwestern.edu>
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org. 

Thank you


The information in this e-mail, and any attachments therein, is confidential and for use by the intended addressee only. If this message is received by you in error please do not disseminate or read further. Please reply to the sender that you have received the message in error, then delete the message. Although Catholic Health Services of Long Island attempts to sweep e-mail  and attachments for viruses, it does not guarantee that either are virus-free and accepts no liability for any damage sustained as a result of viruses. Thank you.





------------------------------

Message: 4
Date: Fri, 8 Feb 2013 11:49:04 -0500
From: "Podawiltz, Thomas" <tpodawiltz <@t> lrgh.org>
Subject: [Histonet] RE: Cassette labeling
To: "Margiotta-Watz, Michele" <MMargiotta <@t> bmhmc.org>, "'SimmsColon,
	Janine'"	<Janine.SimmsColon <@t> jmmc.com>,
	"histonet <@t> lists.utsouthwestern.edu"
	<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
	<38667E7FB77ECD4E91BFAEB8D986386324FCE3BE73 <@t> LRGHEXVS1.practice.lrgh.org>
	
Content-Type: text/plain; charset="us-ascii"

That is basically what we do. 

Tom

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Margiotta-Watz, Michele
Sent: Friday, February 08, 2013 7:55 AM
To: 'SimmsColon, Janine'; histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] RE: Cassette labeling

We print out an embedding log and note the # of pieces/block from the Doc's dictation. We are small scale so it works fine for us.

Michele

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of SimmsColon, Janine
Sent: Friday, February 08, 2013 7:45 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Cassette labeling

Good morning all, 

 

I hope all here in the East are prepared for the storm, (considering I am in New England and it is February it is not all that surprising to me yet I am always amazed at the commotion that occurs with every impending snow, but I digress). I know many out there may have cassette labelers so this may not be an issue but for those who still label by hand I am curious as to how you label. At my previous lab besides the case number (S13-accession number), we also indicated how to embed the pieces (= for on edge), and the number of pieces in the cassette. I am just curious as to how common this practice is or if there is any other way you annotate the number of pieces expected in a cassette. Thank you in advance for any feedback you provide as I have received many helpful responses to my previous posts.

 

Janine Simms Colon, BHA, CPhT, HT(ASCP)

Histology/Pathology

Johnson Memorial Hospital

201 Chestnut Hill Road

Stafford Springs, CT 06076

Office: 860-684-8230 ext. 5197

janine.simmscolon <@t> jmmc.com

 


The information contained in this message may be privileged and confidential and protected from disclosure. 
If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify us immediately by replying to the message and deleting it from your computer.

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
DISCLAIMER:

This e-mail and any files transmitted with it are confidential and are intended solely for the use of the individual or entity to which they are addressed.  This communication may contain material protected by the attorney-client privilege.  If you are not the intended recipient or the person responsible for delivering the e-mail to the intended recipient, be advised that you have received this e-mail in error and that any use, dissemination, forwarding, printing, or copying of this e-mail is strictly prohibited.  If you have received this e-mail in error, please immediately notify the sender via return e-mail or call Brookhaven Memorial Hospital Medical Center at (631) 654-7282.

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
THIS MESSAGE IS CONFIDENTIAL.  
This e-mail message and any attachments are proprietary and confidential information intended only for the use of the recipient(s) named above. If you are not the intended recipient, you may not print,distribute, or copy this message or any attachments.  If you have received this communication in error, please notify the sender by return e-mail and delete this message and any attachments from your computer. Any views or opinions expressed are solely those of the author and do not necessarily represent those of LRGHealthcare.




------------------------------

Message: 5
Date: Fri, 8 Feb 2013 12:41:36 -0500
From: Paula Sicurello <patpxs <@t> gmail.com>
Subject: Re: [Histonet] RE: Temperatures
To: Susan.Walzer <@t> hcahealthcare.com
Cc: histonet <@t> lists.utsouthwestern.edu, BSullivan <@t> virtua.org
Message-ID:
	<CAPSjddZoPayAVOZSMSUHAcxF=mu3Y909-9y9LqmXfSu63rUhQQ <@t> mail.gmail.com>
Content-Type: text/plain; charset=ISO-8859-1

We have NIST traceable digital thermometers that can have the min/max
reset.  We have been told that if you reset on Friday and check on
Monday morning the min/max & it is in range-put an OK by the date.

-- 
Paula Sicurello, HTL (ASCP)
Supervisor, Clinical Electron Microscopy Laboratory
Duke University Health System
Rm.#251M, Duke South, Green Zone
Durham, North Carolina 27710
P:  919.684.2091


On Fri, Feb 8, 2013 at 9:30 AM,  <Susan.Walzer <@t> hcahealthcare.com> wrote:
> We have the clinical lab check our friges and freezers on the weekends..everything else is off.
>
> -----Original Message-----
> From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Sullivan, Beatrice
> Sent: Friday, February 08, 2013 7:52 AM
> To: histonet <@t> lists.utsouthwestern.edu
> Subject: [Histonet] Temperatures
>
> I'm looking for a fix to our problem of no temperatures being taken on the weekends. We are closed and this is creating an issue. Our processors are not running until Sunday night but the paraffin in both the processors and embedding center are kept molten. Any help would be greatly appreciated
>
> Beatrice L. Sullivan HT(ASCP)HTL
> Corporate Histology Manager
> Virtua, Voorhees
> 856-247-3144
>
>
> This message, and any included attachments, are from Virtua Health or its related affiliates and is intended only for the addressee(s). The information contained herein is privileged, proprietary or may include confidential information and/or protected patient health information. Any unauthorized review, forwarding, printing, copying, distributing, or otherwise disseminating or taking any action based on such information is strictly prohibited. If you have received this message in error, or have reason to believe you are not authorized to receive it, please delete this message promptly and notify the sender by e-mail with a copy to ISSECURITY <@t> virtua.org.
>
> Thank you
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet



------------------------------

Message: 6
Date: Fri, 8 Feb 2013 17:58:28 +0000
From: Cathy Crumpton <cathy.crumpton <@t> tuality.org>
Subject: [Histonet] PIN4 staining
To: "histonet <@t> lists.utsouthwestern.edu"
	<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
	<AC889E5706400B438D42F8A07AE3AF8B01817083 <@t> EXMBXSRV01.corp.tuality.net>
Content-Type: text/plain; charset="us-ascii"

We use Bond slides and have noticed that occasionally they seem to have an extra thick coating on them that will pick up DAB and turn slides brown.  If the glass itself is turning brown this might be your issue.  Have you tried different slide lots or brands?

Cathy Crumpton HT(ASCP), Lead Histotechnician

Tuality Community Hospital


------------------------------

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

End of Histonet Digest, Vol 111, Issue 13
*****************************************


The information contained in this e-mail message is intended only for the personal and confidential use of the recipient(s) named above. If the reader of this message is not the intended recipient or an agent responsible for delivering it to the intended recipient, you are hereby notified that you have received this document in error and that any review, dissemination, distribution, or copying of this message is strictly prohibited. If you have received this communication in error, please notify us immediately by e-mail, and delete the original message.


More information about the Histonet mailing list