[Histonet] IHC pos. & neg. control question
Amos Brooks
amosbrooks <@t> gmail.com
Fri May 20 16:04:50 CDT 2011
Thanks Liz,
You are absolutly right there, but have you ever noticed some folks
eyes glaze over when you say that they must both be run at the same ug/mL? I
would be lying if I said it never happened to me until I forced myself to
work it out. Once you do though it isn't too bad. Unfortunately people don't
think about this much. They want to put a slide on a machine and walk away
without thinking about what they are doing. I guess I was oversimplifying
the description. Thanks for pointing that out as it is actually an important
aspect to the dilutions.
Amos
On Fri, May 20, 2011 at 4:25 PM, Liz Chlipala <liz <@t> premierlab.com> wrote:
> Amos
>
> Isotype negative controls are based upon protein concentration not
> dilution. They must be the same protein concentration of the primary
> antibody at the dilution you are using. Using the same dilution will
> not work unless both stock solutions (primary antibody and isotype
> control) are at the same protein concentration which is rarely the case.
>
> Liz
>
> Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
> Manager
> Premier Laboratory, LLC
> PO Box 18592
> Boulder, Colorado 80308
> office (303) 682-3949
> fax (303) 682-9060
> www.premierlab.com
>
>
> Ship to Address:
> 1567 Skyway Drive, Unit E
> Longmont, Colorado 80504
>
> -----Original Message-----
> From: histonet-bounces <@t> lists.utsouthwestern.edu
> [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Amos
> Brooks
> Sent: Friday, May 20, 2011 2:16 PM
> To: jm.lapointe <@t> accellab.com; histonet <@t> lists.utsouthwestern.edu
> Subject: [Histonet] IHC pos. & neg. control question
>
> Hi,
> This is simply a question of definitions. They are actually both
> right.
> If you have the patient tissue as a negative control you are not using
> the
> primary antibody on it but are replacing it (ideally) with an Ig with
> the
> same isotype AND dilution (universal negative is stupid). So if your
> primary
> is an IgG1 from a mouse and you use it at 1:100, your negative control
> would
> be a non-immune mouse serum IgG1 diluted to 1:100. Running this on
> unrelated
> material will not show you anything in this case. what it will show you
> is
> that the detection system you are using is specific to the antibody you
> put
> on the tissue and not to A) something else in the tissue or B) something
> on
> the Ig that is not the epitope itself.
> Although, If you run the same primary antibody on tissue that you
> expect will not react with the primary antibody, this would prove that
> the
> antibody reaction is specific to the antigenin question. This does raise
> the
> question of what to do about ubiquitous antigens (like ubiquitin) that
> are
> actually everywhere. Every cell has a cell cycle so they will all at one
> point or another show proliferation or degeneration for example. Does
> this
> mean you don't run a negative control? No you need to look at it in the
> perspective of expected results.
> There are various ways of using negative (and positive) controls.
> It
> would actually be cost prohibitive to run ALL the possible positive and
> negative controls for every test that we as histologists do. We need to
> discuss with our pathologists what they want to have done to give them
> the
> confidence in our testing to support their diagnosis. But remember that
> just
> because another lab and another pathologist does it differently doesn't
> mean
> it is wrong. It's just different.
>
> Happy Friday Folks,
> Amos
>
>
> Message: 5
> Date: Thu, 19 May 2011 13:25:49 -0400
> From: "Jean-Martin Lapointe" <jm.lapointe <@t> accellab.com>
> Subject: [Histonet] IHC pos. & neg. control question
> To: <histonet <@t> lists.utsouthwestern.edu>
> Message-ID:
> <
> BEFD613BD39142499989F836556DDC830127283C <@t> ACE.accellab.lan>
> Content-Type: text/plain; charset="iso-8859-1"
>
> Hi Curt,
> I agree with your pathologist. The section that you use as a negative
> control (without primary) in an IHC run should ideally be tissue that is
> a
> known positive, and of the same nature as the test sample. Therefore the
> best sample is often a section of your positive control tissue.
> The purpose of this negative control is to make sure that any positive
> staining observed in the test sample is not due to a spurious
> cross-reaction, unrelated to the primary. I don't think the issue per se
> is
> that you are using tissue from the same patient; rather, it is that you
> are
> using a sample for which the staining characteristics are not known. For
> instance, if are using a lymph node section as a negative, for a stain
> that
> targets an epithelial marker (eg Her2) in the test breast sample, then
> your
> negative tissue is not appropriate, because since lymph node contains no
> epithelial tissue, it will not stain no matter what. Therefore if your
> test
> sample shows a positive reaction in the epithelial tissue, but for some
> reason that reaction is a spurious false-positive, then the lymph node
> negative will not reveal that.
> I realize that this is all very theoretical and hypothetical, but I
> understand the pathologist wanting to be confident in the knowledge that
> all
> potential technical issues are eliminated before making his diagnosis.
>
> Jean-Martin
> _______________________________________________