[Histonet] negative controls
Rene J Buesa
rjbuesa <@t> yahoo.com
Fri Oct 15 10:33:37 CDT 2010
Because each tissue block has its own characteristics regarding fixation and processing some of which can influence the reactivity. If you have a bank of negative controls, how can you be sure that any of those blocks have received exactly the same treatment and reacted in the same way to the test block?
The same goes for any bank of positives, so that is why you should have a positive control section in the same slide as the test section.
René J.
--- On Fri, 10/15/10, sgoebel <@t> xbiotech.com <sgoebel <@t> xbiotech.com> wrote:
From: sgoebel <@t> xbiotech.com <sgoebel <@t> xbiotech.com>
Subject: RE: [Histonet] negative controls
To: "Sebree Linda A" <LSebree <@t> uwhealth.org>
Cc: "Histo Net list server" <HistoNet <@t> lists.utsouthwestern.edu>
Date: Friday, October 15, 2010, 11:17 AM
Why do you need a negative control for each block if you are runn= ing
the same antibody on each patient block? Is it just for case by c ase reference so the negative is filed with the patient slide? Why
co= uldn't you have a control slide bank that was dated so all the
slides you d= id on that day, on that run, could be referenced back to
that control? = ; Just curious?
Sarah Goebel, B.A., HT (ASCP)
Histotechnician<= br>
XBiotech USA Inc.
8201 East Riverside Dr. Bld= g 4 Suite 100
Austin, Texas 78744
=
(512)386-= 5107
-------- Original Message --------
Subject: RE: [Histonet] negative controls
From: "Sebree Linda A" <[1]LSebree@= uwhealth.org>
Date: Fri, October 15, 2010 8:08 am
To: "Victoria Baker" <[2]bakevict= oria <@t> gmail.com>, "Histo Net list
server"
<[3]HistoNet <@t> lists.uts= outhwestern.edu>
We run negative controls on every block of a case within the same run.
On autopsy cases, we only run 1 negative per tissue type, within the
same run...this is the only exception to the rule of 1 negative per
block.
Linda A. Sebree
University of Wisconsin Hospital & Clinics
IHC/ISH Laboratory
DB1-223 VAH
600 Highland Ave.
Madison, WI 53792
(608)265-6596
-----Original Message-----
From: [4]histonet= -bounces <@t> lists.utsouthwestern.edu
[[5]mailto:histon= et-bounces <@t> lists.utsouthwestern.edu] On Behalf Of
Victoria
Baker
Sent: Friday, October 15, 2010 9:26 AM
To: Histo Net list server
Subject: [Histonet] negative controls
Hi
I have a hypothetical question to those who run IHC on Ventana
instruments.
Are you running your negatives with your patient/test cases or on a
separate
run? Also, if you are doing this and have to use a different detection
kit
how do you work the QA/QC portion of this for CAP requirements.
Thanks
Vikki
_______________________________________________
Histonet mailing list
[6]Histonet <@t> lists.utsouth= western.edu
[7]http:= //lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
[8]Histonet <@t> lists.utsouth= western.edu
[9]http:= //lists.utsouthwestern.edu/mailman/listinfo/histonet
References
1. 3D"mailto:LSebree <@t> uwhealth.org"
2. 3D"mailto:bakevictoria <@t> gmail.com"
3. 3D"mailto:HistoNet <@t> lists.utsouthwestern.edu"
4. 3D"mailto:histonet-bounces <@t> lists.utsouthwestern.edu"
5. 3D"mailto:histonet-bounces <@t> lists.utsouthwestern.edu"
6. 3D"mailto:Histonet <@t> lists.utsouthwestern.edu"
7. 3D"http://lists.utsouthwestern.edu/mailman/listinfo/histonet"
8. 3D"mailto:Histonet <@t> lists.utsouthwestern.edu"
9. 3D"http://lists.utsouthwestern.edu/mailman/listinfo/histonet"
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list