Pretty sure you have to run your positive with your negative to keep the
conditions 100% the same.  You don't have to put them on the same slide
necessarily, but they need to be on the same run.  Also, say you are
running 15 HP patient slides, you can have one negative control for all
of these as long as it is on the same run.

Sarah Goebel, B.A., HT (ASCP)
Histotechnician


XBiotech USA Inc.

8201 East Riverside Dr. Bldg 4 Suite 100

Austin, Texas  78744

(512)386-5107




-------- Original Message --------
Subject: [Histonet] negative controls
From: Victoria Baker <bakevictoria <@t> gmail.com>
Date: Fri, October 15, 2010 7:25 am
To: Histo Net list server <HistoNet <@t> lists.utsouthwestern.edu>

Hi
I have a hypothetical question to those who run IHC on Ventana
instruments.
Are you running your negatives with your patient/test cases or on a
separate
run? Also, if you are doing this and have to use a different detection
kit
how do you work the QA/QC portion of this for CAP requirements.

Thanks

Vikki