Hi Ashley,
the clearance is between 1-3 mm and I think that's too small. In a few days
Ventana will lift the arm with the hammer for additional 3 mm. That should
help for a reliable working of the instrument.
Half an inch is about 12 mm. I think, not even with the detection-kit-preps
this clearance is reached on our instrument.
What was astonishing for me is, that they never heard about such problems.
Gudrun
-----Ursprüngliche Nachricht-----
Von: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] Im Auftrag von Troutman,
Kenneth A
Gesendet: Freitag, 12. November 2010 19:34
An: 'Histonet <@t> lists.utsouthwestern.edu'
Betreff: RE: [Histonet] question to ventana benchmark XT and ultra
You might not have seated the top of the prep kit container all the way
down. They should not be different heights. I would take each of them and
press down on the top to make sure the lid is on properly. If you set them
side by side and they are still different heights, I would contact your
Ventana rep immediately and show them this. Otherwise, I am inclined to
agree with Clare that the reagent trays were not in the proper position.
You should have about a half inch of clearance between the hammer and the
top of the prep kit container.
Ashley Troutman BS, HT(ASCP) QIHC
Immunohistochemistry Supervisor
Vanderbilt University Histopathology
1301 Medical Center Drive TVC 4531
Nashville, TN 37232
Message: 1
Date: Fri, 12 Nov 2010 11:41:25 -0500
From: Clare Thornton <CThornton <@t> dahlchase.com>
Subject: RE: [Histonet] question to ventana benchmark XT and ultra
users
To: "'Langenberg, Stacey'" <STACEY.LANGENBERG <@t> UCDENVER.EDU>,
"gu.lang <@t> gmx.at" <gu.lang <@t> gmx.at>,
"histonet <@t> lists.utsouthwestern.edu"
<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
<C9D78FFC9D668B4CBEA4405F84697504F801D1DBB6 <@t> iris.dahlchase.net>
Content-Type: text/plain; charset="us-ascii"
Make sure that the lids on your prep-kit dispensers are pushed all the way
down. Since the space between the hammer and the tops of the dispensers is
so small, if one lid is not pushed down all the way the hammer will grab it.
Seen it happen many times myself!
Clare J. Thornton, HTL(ASCP),QIHC
Assistant Histology Supervisor
Dahl-Chase Diagnostic Services
417 State Street, Suite 540
Bangor, ME 04401
cthornton <@t> dahlchase.com
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Langenberg,
Stacey
Sent: Friday, November 12, 2010 11:22 AM
To: gu.lang <@t> gmx.at; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] question to ventana benchmark XT and ultra users
We have had the ultras for over a year now and that is a new one. Was the
rack of antibodies not seated properly? Just a thought.
Stacey
Sent via BlackBerry from T-Mobile
-----Original Message-----
From: Gudrun Lang <gu.lang <@t> gmx.at>
Sender: "histonet-bounces <@t> lists.utsouthwestern.edu"
<histonet-bounces <@t> lists.utsouthwestern.edu>
Date: Fri, 12 Nov 2010 09:12:52
To: histonet <@t> lists.utsouthwestern.edu<histonet <@t> lists.utsouthwestern.edu>
Reply-To: "gu.lang <@t> gmx.at" <gu.lang <@t> gmx.at>
Subject: [Histonet] question to ventana benchmark XT and ultra users
Hi!
This week we've got the Ultra. In the first run it happend, that the
"hammer" touched a Prep-Kit from the side while rotating and pulled the
whole reagens-rack down. (That makes a very uncomfortable noise.)
We saw, that the distance between the hammer and the Prep-Kits is very
small, and it differs from Prep-Kit to Prep-Kit.
Has anyone of the Ventana Users similar experiences?
Bye
Gudrun Lang