[Histonet] Special stains

WILLIAM DESALVO wdesalvo.cac <@t> hotmail.com
Tue Jan 5 20:11:16 CST 2010


I agree with the majority that you MUST do a validation of the instrument before you run any patient diagnostic samples. Any time you make a change in instrumentation, reagents, procedure or protocol you should run at least a quick validation study to compare results of the new variable(s) to your control method (current) and it must be reviewed and approved by the medical director of your lab. The numbr of samples is always the question and I suggest you investigate whether or not your institution has a an expectation for sample size. The Clinicl lab will be a good place to get information on procedure and process for validation.

That said and for an automated stainer in the Histology lab (Dako or Ventana), I suggest a minimum of three samples on separate runs for each special stain you will perform. If all three are comarable and correlate to the contol method, you can have confidence that you can make the switch and start turnung out patient samples. If you do not create comparable results for the first three, then consult w/ your pathologist as to how many more samples he will feel comfortable reviewing to confirm the performance. You should always stive to produce >95% concordance. I also suggest that you use the "blind study" technique when presenting to the reviewing pathologist. You do not want perception or predjudice to skew your review results. Of course, document all that you do and create a validation folder to house the documents for easy review.

I believe setting up a validation process/protocol for your lab will create more confidence in the instrument, the reagents, the technicians performing the work and you. I have found it is easier to be proactive rather than reactive in producing quality results. Good luck with moving to the automated special stainer. I have used the DAKO for 4+ years and find it will produce a quality stain consistantly.


William DeSalvo, B.S., HTL(ASCP)
Chairperson, NSH QCC




> Date: Tue, 5 Jan 2010 11:33:35 -0800
> From: KGroeger <@t> USLABS.net
> To: Kim.Donadio <@t> bhcpns.org; rjbuesa <@t> yahoo.com
> Subject: RE: [Histonet] Special stains
> CC: histonet <@t> lists.utsouthwestern.edu; Dorothy.L.Webb <@t> HealthPartners.Com; histonet-bounces <@t> lists.utsouthwestern.edu
> 
> I agree, we are validating some new instruments now and have found a lot of variables.
> 
> Karin Groeger
> 
> Histology Supervisor
> 
> US LABS, Irvine,CA
> 
> 949-450-0145 ext. 649
> 
> -----Original Message-----
> From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Kim.Donadio <@t> bhcpns.org
> Sent: Tuesday, January 05, 2010 10:28 AM
> To: Rene J Buesa
> Cc: 'histonet <@t> lists.utsouthwestern.edu'; histonet-bounces <@t> lists.utsouthwestern.edu; Dorothy LWebb
> Subject: Re: [Histonet] Special stains
> 
> In the end it is still a new instrument. It should be validated for all 
> the stains you do on it. This would seem like best practice to me. 
> 
> 
> 
> Kim Donadio 
> Pathology Supervisor
> Baptist Hospital
> 1000 W Moreno St.
> Pensacola FL 32501
> Phone (850) 469-7718
> Fax (850) 434-4996
> 
> 
> 
> Rene J Buesa <rjbuesa <@t> yahoo.com> 
> Sent by: histonet-bounces <@t> lists.utsouthwestern.edu
> 01/05/2010 11:38 AM
> 
> To
> "'histonet <@t> lists.utsouthwestern.edu'" <histonet <@t> lists.utsouthwestern.edu>, 
> Dorothy LWebb <Dorothy.L.Webb <@t> HealthPartners.Com>
> cc
> 
> Subject
> Re: [Histonet] Special stains
> 
> 
> 
> 
> 
> 
> Dorothy:
> I would NOT do it.
> If you do not change the procedure it does not really matter which auto 
> stainer you use. Your "validation" is the reaction of your control that 
> should react as before.
> As a matter of fact, I would be more inclined to made a validation if you 
> change your control rather than if you change the auto stainer.
> If, for example, you are doing IHC and your dilution has been "fine tuned" 
> for a given tissue control and you change the tissue, you will need to 
> "fine tune" the dilution to have the same reaction strength of the epitope 
> in the new control, but you will not have to change your procedure. That 
> is why I always used IHC internal controls, and by using "normal strength 
> epitopes" I don't have to keep validating.
> That is also why I always shied away from using pathological cases to use 
> as epitope controls.
> So, I would not advise you to go through validation because of changing 
> the auto stainer, as long as the procedure is the same.
> René J.
> 
> --- On Tue, 1/5/10, Webb, Dorothy L <Dorothy.L.Webb <@t> HealthPartners.Com> 
> wrote:
> 
> 
> From: Webb, Dorothy L <Dorothy.L.Webb <@t> HealthPartners.Com>
> Subject: [Histonet] Special stains
> To: "'histonet <@t> lists.utsouthwestern.edu'" 
> <histonet <@t> lists.utsouthwestern.edu>
> Date: Tuesday, January 5, 2010, 12:23 PM
> 
> 
> We are switching automated special stain equipment to a different vendor 
> and am wondering if we have to validate all of the special stains we do.  
> I am from the "old school" and the validation process, other than IHC, is 
> not my forte!!  Any advice would be appreciated, so, thanks ahead of 
> time!!
> 
> Dorothy Webb, Ht (ASCP)
> Regions Histology Technical Supervisor
> 651-254-2962
> 
> 
> 
>   ________________________________
> This e-mail and any files transmitted with it are confidential and are 
> intended solely for the use of the individual or entity to whom they are 
> addressed. If you are not the intended recipient or the individual 
> responsible for delivering the e-mail to the intended recipient, please be 
> advised that you have received this e-mail in error and that any use, 
> dissemination, forwarding, printing, or copying of this e-mail is strictly 
> prohibited.
> 
> If you have received this e-mail in error, please immediately notify the 
> HealthPartners Support Center by telephone at (952) 967-6600. You will be 
> reimbursed for reasonable costs incurred in notifying us. HealthPartners 
> R001.0
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 
> 
> 
>  
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 
> 
> 
> -----------------------------------------
> All electronic data transmissions originating from or sent to
> Baptist Health Care Corporation (BHC) are subject to monitoring.
> This message along with any attached data, are the confidential and
> proprietary communications of BHC and are intended to be received
> only by the individual or individuals to whom the message has been
> addressed. If the reader of this message is not the intended
> recipient, please take notice that any use, copying, printing,
> forwarding or distribution of this message, in any form, is
> strictly prohibited and may violate State or Federal Law. If you
> have received this transmission in error, please delete or destroy
> all copies of this message.  For questions, contact the BHC Privacy
> Officer at (850) 434-4472.  Rev.10/07.
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
> 
> 
> 
> Confidentiality Notice: This message, including any attachments, may contain CONFIDENTIAL AND/OR LEGALLY PRIVILEGED information. The information is intended for use by the individual named above and may not be disseminated to any other party without US LABS' written permission. If you are not the intended recipient, or the employee or agent responsible for delivering this information to the intended recipient, you are hereby notified that any dissemination, disclosure, distribution, copying or taking of any action in reliance on the contents of this information is strictly prohibited. If you have received this information in error, please notify US LABS immediately at 1-888-450-0145 attn: Compliance Department to arrange for return of this message including all attachments.
> 
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 		 	   		  
_________________________________________________________________
Hotmail: Powerful Free email with security by Microsoft.
http://clk.atdmt.com/GBL/go/171222986/direct/01/


More information about the Histonet mailing list