[Histonet] LabVision/Dako Autostainer

CHRISTIE GOWAN christiegowan <@t> msn.com
Fri Jan 23 14:50:08 CST 2009


Hey Vinnie,
We have 2 lab visions and we also experienced this problem. They came in and changed the level sensors and we have not had the problem since. 
Christie
UAB Hospital
Birmingham, AL> From: dellav <@t> musc.edu> To: histonet <@t> lists.utsouthwestern.edu> Date: Fri, 23 Jan 2009 15:13:36 -0500> Subject: [Histonet] LabVision/Dako Autostainer > > We have earlier generation Dako Autostainers and are evaluating a current generation LabVision stainer. New or old we have experienced occasions when this instrument design has failed to dispense reagent. This seems to occur randomly or if there is a pattern we haven't figured it out. I've recently learned that some owners of the Biocare Nemesis which is essentially the same instrument have also encountered this problem.> > The arm will travel to the correct slide but nothing dispenses from the probe, then the arm moves on its way. > > I'd appreciate learning if others have experienced this problem and if a solution to the problem was discovered I hope you'll share it with me.> > > > Vinnie Della Speranza> Manager for Anatomic Pathology Services> Medical University of South Carolina> 165 Ashley Avenue Suite 309> Charleston, South Carolina 29425> Tel: (843) 792-6353> Fax: (843) 792-8974> > > > _______________________________________________> Histonet mailing list> Histonet <@t> lists.utsouthwestern.edu> http://lists.utsouthwestern.edu/mailman/listinfo/histonet


More information about the Histonet mailing list