[Histonet] Cloudy MMA embedded block
Pamela Marcum
mucram11 <@t> comcast.net
Tue Apr 21 08:45:05 CDT 2009
Hi,
We don't store our MMA or DBP in the refirgerator as the movement between 4C and room temperature can cause condensation in the bottles. This just adds water to the material and as it is cooled and re-warmed you are adding more each time and in both directions. When it comes to room temperature and then is replaced in the refrigerator it again will develop condensation. Over time the MMA will have enough water in it to become discolored. I am surprised you are not seeing the discoloration in the MMA liquid as it poured as a warning not to use it.
We have fair control of temperature in the laboratory and are generally between 68F to 75F with the average as around 70F. We have never had a problem with the MMA polymerizing or developing a water content. It is stored in a dark cabinet that is in an area I would say is a little cooler than the room. We generally don't polymerize in the cold at 4C however, you timing seems long for the process. It is best not to disturb the vials too much once you are in the embedding phase. Movement or inversion of the vials can cause a problem as the MMA will polymerize from the bottom up and remixing it can slow the process. Also make sure you tops are tight at this step.
Pam Marcum
UPENN Vet Sch
New Botlon Center
----- Original Message -----
From: "Jack Ratliff" <ratliffjack <@t> hotmail.com>
To: "Ooi Ting Huay" <ooi.ting.huay <@t> nhc.com.sg>, "<histonet <@t> lists.utsouthwestern.edu> <histonet <@t> lists.utsouthwestern.edu>" <histonet <@t> lists.utsouthwestern.edu>
Sent: Tuesday, April 21, 2009 9:24:29 AM GMT -05:00 US/Canada Eastern
Subject: Re: [Histonet] Cloudy MMA embedded block
Ooi,
Somehow you are getting moisture/water mixed in with the MMA
solution(s). I only remember visiting the IMCB facility, so if this is
not your facility then I am not sure of your laboratory environment
(air temperature/humidity). The best thing I can tell you at this
point is to try and find ways to reduce the chance for moisture
accumulation.
I am pretty sure that at a minimum you are storing your catalyst and
monomer (MMA) at 4C and your softner (DBP) at room temp. Maybe you
should take extra steps to ensure your chemicals are at room temp
before any use. Lastly, your polymerization is taking too long for
your small 20mL vials. My specimens of similar size require only 3-5
days to completely polymerize. Now, I dont remember your concentration
of catalyst, but if it is around 2.5g/1L embedding solution, try
skipping the 4C step and use room temp (<23C) until the specimens are
polymerized. Then finish them off in the oven when they are tacky or
mostly hard on top. The 4C step is mostly to slow the polymerization
for larger specimens requiring a larger volume of solution and/or when
the lab temp cannot be controlled routinely at or below 23C. The water
bath then also helps to regulate the temperature of the polymerization
in the case where the lab environment is too cool (<20C) and retarding
the rate of polymerzation.
The main thing to keep in mind is that the lab environment pretty much
controls the process and is probably the main variable that will
contrast labs that perform the exact same method.
Jack
On Apr 21, 2009, at 12:33 AM, ooi.ting.huay <@t> nhc.com.sg wrote:
> Hi,
>
> I am trying very hard to get a clear MMA embedded block. I am
> appreciate
> if there is any advice on doing it. I am using 100% MMA (liquid) and
> 0.5%
> Perkadox 16 (powder) to do the embedding. I have tried to embed the
> sample
> in different conditions. However, it seems to be extremely hard to
> get a
> clear embedded block.
>
> Below are the methods that I have tried.
>
> 1. Before embedding, bring the MMA and perkadox 16 to room
> temperature.
> After a few mix it by invertion, I aliquot the solution to the glass
> vials
> that contain samples or simply glass vials that without sample. I
> put the
> glass vials (both with and without sample) in 4 degree, room
> temperature
> and even 37degree oven.
>
> 2. Mix the MMA and perkadox when it is cold, invert it a few times.
> Aliquot it to glass vials that contain sample or without sample. Put
> in
> 4degree, room temperature and 37degree oven.
>
> I have also tried to speed up the polymerization process by using
> vacumm
> for a few samples...
>
> Throughout all the sample that I have, I only manage to get a clear
> empty
> glass vial which I put in the 4degree for one month and a clear
> embedded
> sample which I put in 4 degree overnight and bring it out to room
> temperature. The rest of the glass vials (with or without sample) are
> cloudy.
>
> I have repeated the methods that I used to get a clear embedded
> block, I
> did not get any clear block (for both with and without sample).
>
> I am glad if there is any expert advice or guidance. We need a clear
> embedded sample. Thank you very much!
>
>
>
> Regards,
> Ooi
> _________________________________________________
> Confidential information. Unauthorized use or disclosure prohibited.
> Refer http://www.singhealth.com.sg/ContactUs/#Disclaimer
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list