It depends upon your SOP's most calibrated thermometers have a little
statement at the bottom of the certification that states that most
institutions recalibrate once a year but you must follows your own SOP's.
You then need to take into consideration what guidelines you are following
and what the standards are.  What I'm saying is that you can write in your
SOP's that you are going to calibrate once every 5 years, but when you get
inspected will the inspector think that is good enough. The kicker is the
cost to recalibrate is more than a new thermometer, we just purchase a new
calibrated thermometer each year and that's what is in our SOP's.

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Manager
Premier Laboratory, LLC
P.O. Box 18592
Boulder, CO 80308
phone (303) 735-5001
fax (303) 735-3540
liz <@t> premierlab.com
www.premierlab.com
 
Ship to Address:
 
Premier Laboratory, LLC
University of Colorado at Boulder
MCDB, Room A3B40
Boulder, CO 80309

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Lester Raff
Sent: Thursday, February 15, 2007 10:31 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Thermometers

Do labs generally recalibrate/recertify NIST calibrated thermometers?
If so, how and how often?

 

Lester J. Raff, MD
Medical Director
UroPartners, LLC Laboratory

2225 Enterprise Dr. Suite 2511

Westchester, IL 60154


ph:  708-486-0076
fax: 708-486-0080

 

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


__________ NOD32 1971 (20070110) Information __________

This message was checked by NOD32 antivirus system.
http://www.eset.com





More information about the Histonet mailing list