[Histonet] HELP
mari.ann.mailhiot <@t> leica-microsystems.com
mari.ann.mailhiot <@t> leica-microsystems.com
Tue Aug 21 17:15:21 CDT 2007
HI Dorothy
It sounds like xylene is in the wax. The other possibility could be water.
Check the bottom of the paraffin tank and see if you have droplets. If you
do then it is more than likely water.
Regards
Mari Ann Mailhiot BA HT ASCP
Application Specialist
Leica Technical Assistance Center
800 248 0123 x7267
847 236 3063 fax
mari.ann.mailhiot <@t> leica-microsystems.com
www.leica-microsystems.com
"Webb, Dorothy L"
<Dorothy.L.Webb <@t> H
ealthPartners.Com To
> histonet <@t> lists.utsouthwestern.edu
Sent by: cc
histonet-bounces@
lists.utsouthwest Subject
ern.edu [Histonet] HELP
08/21/2007 12:19
PM
What would cause the tissues when taken out of the processor at the end
of the cycle to feel "slippery", almost greasy? I was off and back
today and when the tech took the cassettes out of the processor, she
mentioned this and they are not cutting well at all. Our processor was
just changed yesterday and as far as I can tell from smell, all reagents
are in order! Where should I start????? Thanks so much for answering
this so quickly..I am stymied!!
Dorothy Webb, HT (ASCP)
Histology Technical Supervisor
Regions Hospital, Pathology Department
640 Jackson Street, Saint Paul, MN 55101-2595
Phone: 651-254-2962
Fax: 651-254-2741
Regions Hospital is part of the HealthPartners family of care
________________________________________
This e-mail and any files transmitted with it are confidential and are
intended solely for the use of the individual or entity to whom they are
addressed. If you are not the intended recipient or the individual
responsible for delivering the e-mail to the intended recipient, please be
advised that you have received this e-mail in error and that any use,
dissemination, forwarding, printing, or copying of this e-mail is strictly
prohibited.
If you have received this e-mail in error, please immediately notify the
HealthPartners Support Center by telephone at (952) 967-6600. You will be
reimbursed for reasonable costs incurred in notifying us.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email
______________________________________________________________________
More information about the Histonet
mailing list