Yes, I've been told, that it works. But what is the histochemical background? 
Proper fixed tissue would'nt suffer too much from high temperatures, I think. What is about too thick, underfixed tissue? Wouldn't it become hard and tough, difficult to cut? What about shrinkage?
What about collagen-rich tissue? I don't want to produce glue.
Is the whole process sensitiv to the duration in the high temp.? What is the recommandation for the maximum duration in 80 degrees?
What are the critical points?

Many questions, I know. But I don't like black boxes. 

Gudrun Lang
 
Biomed. Analytikerin
Histolabor
Akh Linz
Krankenhausstr. 9
4020 Linz
+43(0)732/7806-6754

-----Ursprüngliche Nachricht-----
Von: Phil McArdle [mailto:pmcardle <@t> ebsciences.com] 
Gesendet: Montag, 02. April 2007 20:30
An: gu.lang <@t> gmx.at
Cc: histonet <@t> lists.utsouthwestern.edu
Betreff: Re: AW: [Histonet] Peloris Processor and isopropanol

I'm with a laboratory microwave vendor:

Never mind 74˚ C - some microwave protocols call for temperatures as 
high as 82˚ or 84˚ C (!) in order to exceed the boiling point of 
isopropanol, facilitating its replacement with paraffin. Naturally, this 
caused some concern and resistance in the anatomic pathology community. 
However, look at your own question about IHC: when you consider antigen 
retrieval protocols call for 100˚C, even 84˚ pales by comparison. Of 
course, there are many reasons you want to minimize the amount of time 
spent at high temperatures, but there is no problem with the majority of 
tissue types for the relatively short timeframes involved in microwave 
processing.

Also, by the paraffin step, the tissue has already been fixed, 
dehydrated, and defatted; while it is in a sense still "tissue," it's 
radically different from the tissue as received. You certainly don’t 
want to burn or "cook" anything, but 82°C isn’t really extreme in this 
context. (I'd be at least as concerned about how a given paraffin and 
additives hold up at high temperatures, so it's always a good idea to 
check with the manufacturer.) There also is the issue of vacuum; since 
the purpose of high temperature is to get above the boiling point of 
isopropanol, and vacuum causes BP depression, vacuum should allow you to 
lower the temperature to something you're more comfortable with.

More information may be found at

http://www.ebsciences.com/pdf/EBS_MW_COMPANION.pdf

Very best regards,

Phil McArdle

-- 
Phil McArdle
Microwave Product Manager

Energy Beam Sciences, Inc.
29-B Kripes Rd.
East Granby, CT 06026

Tel:  800.992.9037 x 341
Mobile: 860.597.6796
Fax: 860.653.0422

pmcardle <@t> ebsciences.com
www.ebsciences.com

You must be the change you want to see in the world.
- Mahatma Gandhi

NOTE: This message, together with any attachments, is intended only for the use of the individual or entity to which it is addressed and may contain information that is legally privileged, confidential and exempt from disclosure. If you are not the intended recipient, however, there's not a lot I can do about it, and it was probably my mistake anyway. So please do the right thing and make this e-mail go away. Thank you.



Gudrun Lang wrote:
> I've seen a protocol for isopropanol-processing with an IPA-temperature
> about 74 degrees. That is near the boilingpoint and means rather a boiling
> out of the tissue than acting as intermedium. Similar protocols are found
> with the microwave-techniques. I think there is no principal difference in
> reaching the temperature.
>
> What I'm a little confused about is the fact, that usually recommended
> temperatures for tissueprocessing don't raise above 62 degrees. So how does
> this influence staining, ihc-staining etc.?
> What are the temperatures in your protocol?
> What does the community think about this issue?
>
> Gudrun Lang
>  
> Biomed. Analytikerin
> Histolabor
> Akh Linz
> Krankenhausstr. 9
> 4020 Linz
> +43(0)732/7806-6754
>
> -----Ursprüngliche Nachricht-----
> Von: histonet-bounces <@t> lists.utsouthwestern.edu
> [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] Im Auftrag von Judy
> Collins
> Gesendet: Montag, 02. April 2007 15:53
> An: histonet <@t> lists.utsouthwestern.edu
> Betreff: [Histonet] Peloris Processor
>
> We have the Peloris rapid tissue processor.  It is xylene free but does
> not use microwaves.  It uses isopropyl alcohol after the reagent alcohol
> to remove all traces of water and then goes straight to the paraffin
> >from there.  Amazingly, and contrary to everything we have been taught,
> this works!
>
> It processes small biopsies in approximately 1 hour.  Shave biopsies
> take approximately 2 hours and the larger specimens 4 to 6 hours.  We
> find the processing to be at least equal to routine processing and in
> many cases, superior to it.  Because it does not use xylene, the tissues
> are not hard and brittle as they can sometimes be with routine
> processing.
>
> It has two separate retorts which can run two separate runs at the same
> time.  We are able to process our specimens throughout the day instead
> of everything being processed at night as in the past. It has greatly
> improved our TAT.
>
> The Peloris tracks the number of blocks run and, based on this, tells
> you when to change the various solutions.  As a result, we have saved a
> significant amount of money on reagents. 
>
> We are extremely happy with ours.
>
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>   




More information about the Histonet mailing list