[Histonet] Leica microtome
RCHIOVETTI <@t> aol.com
RCHIOVETTI <@t> aol.com
Fri Jul 14 14:59:24 CDT 2006
In a message dated 7/14/2006 12:36:59 PM US Mountain Standard Time,
ftryka <@t> tetonhospital.org writes:
> We are seeking advice. Our Leica 2125 RM microtome has been
> functioning well until it suddenly started cutting thick thin sections.
> We have tried to align it, etc and it still is not functioning properly.
>
> The block face looks smooth, but it will cut several slices toward one
> side, then on the next pass it cuts off a potato chip slice. Any
> suggestions would be appreciated. Franci Tryka
>
>
Franci,
Most of the time when microtomes start acting up like this, there's a problem
either with the clamping of the blade in the blade holder or with the knife
holder itself.
I'd check for paraffin or other debris between the clamping plates, a nick or
a defect on the front or back clamping plate, the clamping pressure of the
blade or the tightness of the knife holder where it clamps onto the base of the
microtome.
If all of that fails, you can contact Leica at 1-800-248-0123.
Hope this helps!
Cheers,
Bob
Robert (Bob) Chiovetti, Ph.D.
Owner
The Microscope Works
Arizona's Microscopy Resource
132 North Elster Drive
Tucson, AZ 85710-3212 USA
Tel./Fax 520-546-4986
Manufacturers' Representative
Member, Arizona Small Business Association - ASBA
More information about the Histonet
mailing list