[Histonet] cleaning microtomes
RCHIOVETTI <@t> aol.com
RCHIOVETTI <@t> aol.com
Tue Sep 9 13:15:18 CDT 2003
In a message dated 09/09/2003 7:51:39 AM US Mountain Standard Time,
Jmrwalker <@t> aol.com writes:
<< After you are finished cutting on
your microtome, what do you use to clean it? Our lab in the past always
used
xylene or alcohol to clean it. A few new people have been hired in the lab
and
now Paragard and baby oil have been used. I'm just curious what the
majority
of people in histoland are using. Thanks! >>
I see mostly Paragard being used in our customers' labs, and I agree with the
other posters, if the "slick" surface is a problem after the Paragard
treatment you can then wipe down with a little ethanol or isopropanol.
With time xylene can affect the surfaces of the microtome's external parts,
especially if used in excess. Lots of the microtome parts have anodized
(plated) surfaces that are either black or clear, and you'd think they would hold up
to just about any kind of abuse. But they will definitely break down and
corrode with prolonged or repeated exposure to xylene. This can lead to pitting
and discoloration of the surfaces.
If you *do* use xylene, use it sparingly and be sure to wear gloves. I
personally don't like the stuff. Try to use a substitute instead.
First, use a brush or gauze pad to remove the majority of the paraffin. Then
use only enough xylene or xylene substitute to barely moisten a gauze 4x4 or
a tissue. Wipe off the excess immediately with another dry gauze or tissue.
You can also follow this with an alcohol wipe-down like you would use with
Paragard if the surface feels slick or oily after you clean it.
Hope this helps!
Bob Chiovetti
GTI Microsystems
Leica Regional Dealer
Desert Southwest Region, USA