<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>A 
cautionary note</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>I 
agree that false positives are rare but they do happen. We had a couple a while 
ago when it turned out that the Teflon coat on the probe of our DAKO Autostainer 
had perished and was carrying over reagents from one slide to another and also 
contaminating our negative control solution with primary 
antibody.</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>Also 
they are very useful in checking that there is no endogenous biotin activity 
(more common than people think) and also that your hydrogen peroxide solution 
hasn't gone off when quenching endogenous peroxidase 
activity.</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>As for 
the argument about only one positive control per antibody per run rather than 
per patient - </SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>Well 
the question I would pose is this. If in 5 years time you wish to review the 
immuno for a particular case, how do you ensure that the staining quality was 
adequate at the time if there aren't any positive controls for that 
case?</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>On 
medico-legal grounds alone you should be doing positive controls for each case, 
especially if those results end up with the patient having a particularly 
aggressive treatment like chemotherapy or a particularly invasive surgical 
procedure such as a colectomy or a mastectomy. If someone wants to sue your 
organisation at a later date for inappropriate treatment you need every bit of 
proof that your part of the investigation was above reproach and controls is one 
thing that will definitely be asked about. I certainly wouldn't like to stand up 
in court and try and defend myself against that one!</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003>Incidentally, we use Surgipath's Control slides which 
allow you (most of the time depending on the size of the test section) to have 
the positive control and test section on the same slide, which saves a lot of 
space on the immunostainer and neatly solves the problem of where do you store 
the positive control if you use the single control per run model and you have 
multiple cases with the same antibody. It also acts as a check that the slide 
has received the correct primary antibody and that someone hasn't loaded the 
immunostainer wrongly.</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>At the 
end of the day I still go by the analogy someone else made here earlier about 
airbags and seat belts in cars. Would you drive in a car without them? They may 
never be needed in your entire driving life, but you would be a dam fool not to 
have&nbsp; them wouldn't you?</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>And 
I&nbsp;think we will have to agree to&nbsp;disagree Hadi, with your last 
statement, it is poor quality control by any definition of the term, not to use 
both negative and positive controls for each case.</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003></SPAN></FONT>&nbsp;</DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN class=812535506-11102003>Nick 
Kirk</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003>Histopathology</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003>Hinchingbrooke Hospital</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003>Huntingdon</SPAN></FONT></DIV>
<DIV><FONT face=Arial color=#0000ff size=2><SPAN 
class=812535506-11102003>England</SPAN></FONT></DIV>
<BLOCKQUOTE dir=ltr style="MARGIN-RIGHT: 0px">
  <DIV class=OutlookMessageHeader dir=ltr align=left><FONT face=Tahoma 
  size=2>-----Original Message-----<BR><B>From:</B> 
  histonet-admin@lists.utsouthwestern.edu 
  [mailto:histonet-admin@lists.utsouthwestern.edu]<B>On Behalf Of </B>Hadi 
  Yaziji<BR><B>Sent:</B> 11 October 2003 03:07<BR><B>To:</B> 
  histonet@pathology.swmed.edu<BR><B>Subject:</B> Re: [Histonet] negative IHC 
  controls<BR><BR></FONT></DIV>Theoretically, I agree with you that (a) running 
  multiple negative control sections to cover the different 
  antibodies/pretreatments is ideal and (b) money should be well spent on QC (I 
  can't emphasize that enough). However, I honestly don't remember when was the 
  last time when I had a false positive staining on a patient's slide that the 
  negative control was the only way that allowed me to recognize the false 
  positive signal on the real slide. And we look at 100s of IHC stains every 
  day. <BR><BR>Experienced pathologists and technologists can still recognize in 
  &gt; 99% of the times a false positive signal from a real signal on the tissue 
  section of a given antibody without even having to look at negative controls. 
  I don't think it should be a CAP requirement at all. In fact, I'd say the same 
  thing on positive controls. All you need is one positive control per antibody, 
  but not one control/per antibody/per patient. In many cases positive internal 
  controls are present on the same slide, so you can tell whether your antibody 
  worked simply by evaluating the expected positive internal controls. In fact, 
  if your positive 'external' control worked and internal controls didn't, then 
  you need to repeat your antibody test regardless of the positive external 
  control. External controls are different tissues, fixed differently, processed 
  differently and the tissue ages differently (in terms of its antigenicity). 
  <BR><BR>Individuals who perform and interpret IHC studies must have the 
  knowledge to recognize the expected sub-cellular localization of every 
  antibody on every type of tissue, including aberrant signals. For instance, 
  you can easily dismiss a granular cytoplasmic TTF-1 signal as not positive, 
  because TTF-1 is a nuclear transcription factor and we know it's expressed in 
  the nuclei. However, hepatocellular carcinomas can give you a consistent and 
  reproducible granular and cytoplasmic TTF-1 signal, and if you subsequently 
  run confirmatory markers (HepPar1, polyclonal CEA, CD10, etc.) they will be 
  positive. Such (granular cytoplasmic) signal should be recognized by the 
  interpreter (tech or pathologist) as a specific signal for tumors with 
  hepatoid phenotypes. This is just one of many examples..<BR><BR>The bottom 
  line: in real life, one negative control per case is <B>more than 
  sufficient</B>. And to say the least, it's inaccurate to state this is poor 
  patient care and lousy quality control.<BR><BR>I look forward to any 
  constructive criticism. <BR><BR>Hadi Yaziji, M.D.<BR>PhenoPath 
  Laboratories<BR><BR>On Thursday, October 9, 2003, at 07:03 AM, Horn, Hazel V 
  wrote:<BR><BR>
  <BLOCKQUOTE><?fontfamily><?param Comic Sans MS><?color><?param 0000,0000,FFFF>Not 
    only should you be running a negative control for each patient 
    slide.&nbsp;&nbsp; That negative control should be treated just as your 
    antibody is.&nbsp;&nbsp; If the antibody is rabbit and antigen retrieved, so 
    should your control.&nbsp;&nbsp; If another antibody on the same patient is 
    mouse and not retrieved another negative control should be run with this 
    same protocol.&nbsp;&nbsp; In the United States, labs that are inspected by 
    the CAP, are required to run these controls.&nbsp;&nbsp;&nbsp; MONEY should 
    never be considered as a reason to stop doing a part of a 
    procedure.&nbsp;&nbsp; It's poor patient care and lousy quality 
    control.&nbsp;&nbsp; 
    IMHO.<?/color><?/fontfamily><BR>&nbsp;<BR><?smaller>Hazel Horn, HT/HTL 
    (ASCP)<BR>Histology Supervisor<BR>Arkansas Children's Hospital<BR><BR>Phone 
    - 501.364.4240<BR>Fax - 501.364.3912<?/smaller><BR><BR><BR><?fontfamily><?param Tahoma><?smaller>-----Original 
    Message-----<BR><B>From:</B> vermast 
    [mailto:vermast@rogers.com]<BR><B>Sent:</B> Wednesday, October 08, 2003 3:57 
    PM<BR><B>To:</B> histonet@lists.utsouthwestern.edu<BR><B>Subject:</B> 
    [Histonet] negative IHC controls<BR><BR><?/smaller><?/fontfamily>&nbsp;<BR><?fontfamily><?param Arial><?smaller>I 
    would like to get a feel for how many out there are running negative control 
    slides for IHC.&nbsp;<?/smaller><?/fontfamily><BR>&nbsp;<BR><?fontfamily><?param Arial><?smaller>In 
    our lab we do just a handful of antibodies and initially I had been running 
    a negative control slide with each patient slide. &nbsp; After much 
    discussion with our pathologists, we decided to omit these negatives (which 
    were conistently negative)&nbsp;and continue to just run a positive control 
    with each primary antibody for the run.&nbsp; We use the Dako autostainer 
    and prediluted primaries.&nbsp; The decision to stop running negatives also 
    coincided with Dako's decision to sell the negative control sera separately 
    from the primaries (they used to come packaged together).&nbsp; Perhaps I 
    assumed that discontinuing to pair these reagents together meant that few 
    labs were using the negatives.<?/smaller><?/fontfamily><BR>&nbsp;<BR><?fontfamily><?param Arial><?smaller>Anyhow, 
    after having reviewed the last QMPLS (Canada) survey committee comments, I 
    believe the committe would like a negative control run with each patient 
    tissue slide in order to evaluate background&nbsp; (they have used NCCLS 
    guide pages as reference).&nbsp; Incidentally we weren't a part of the 
    survey due to a technicality.<?/smaller><?/fontfamily><BR><?fontfamily><?param Arial><?smaller>Any 
    help or advice would be appreciated.<?/smaller><?/fontfamily><BR>&nbsp;<BR><?fontfamily><?param Arial><?smaller>L. 
    Vermast<?/smaller><?/fontfamily><BR><?fontfamily><?param Arial><?smaller>Stratford, 
    Ont.<?/smaller><?/fontfamily><BR><BR><BR><BR><BR><BR></BLOCKQUOTE>&lt;image.tiff&gt;<BR>
  <BLOCKQUOTE><BR>The information contained in this message may be privileged 
    and confidential and protected from disclosure. If the reader of this 
    message is not the intended recipient, or an employee or agent responsible 
    for delivering this message to the intended recipient, you are hereby 
    notified that any dissemination, distribution or copying of this 
    communication is strictly prohibited. If you have received this 
    communication in error, please notify us immediately by replying to the 
    message and deleting it from your computer. Thank you. Arkansas Children's 
    Hospital<BR></BLOCKQUOTE></BLOCKQUOTE></BODY></HTML>