[Histonet] Leak under Sakura Tissue-Tek TEC 5 embedding machine

travis at imebinc.com travis at imebinc.com
Tue Feb 27 12:31:56 CST 2024


Hello,

It's a pretty common issue that shouldn't impact your embedding. The TEC 5 is not air tight, so when its moved with paraffin filled to the rim, some of it spills and falls into the cracks. Over time it builds up inside the unit by melting and getting hard again. Until one day it moves all the way through unit and leaks out at the lowest point (normally under one of the two chambers). When we refurbished these units we open them up to find them filled with paraffin but, it doesn’t really cause any problems other than a mess under the unit. 😊


Travis O’Brien
 
International Medical Equipment, Inc.
170 Vallecitos De Oro
San Marcos, CA 92069
 
800.543.8496  Phone
760.761.0859  Fax
www.imebinc.com

-----Original Message-----
From: M.O. via Histonet <histonet at lists.utsouthwestern.edu> 
Sent: Tuesday, February 27, 2024 9:51 AM
To: Histonet <histonet at lists.utsouthwestern.edu>
Subject: [Histonet] Leak under Sakura Tissue-Tek TEC 5 embedding machine

Hello,

Has anyone encountered a leak of liquid paraffin under their embedding machine? We have a TEC 5 and I'm not sure where to start in trying to locate the leak. If anyone has any insights, that would be great!

Thank you,
Merissa