I am wondering if you are getting air bubbles/pockets in the cassettes themselves.  Are you using mesh cassettes?  Are you loading the trays with cassettes into an empty retort and then starting the delayed program?  If so,  air pockets could be forming in the cassette and around the tissue pieces.  Since there is no p/v cycle running yet, those air pockets may be just sitting there and the exposed tissue pieces may be getting dried out.

I suggest that once the retort is full and in ambient mode, open retort and move the cassette trays around a bit and see if air bubbles are rising from the cassettes.

When the trays of cassettes sit on the bench in formalin waiting to go on the VIP,  there may be enough movement when moving the container, etc,  that any air pockets are disrupted and the tissues are  not exposed to air pockets.

good luck,
amysue ruppert
Marshfield Labs Histology
________________________________
From: histonet-request at lists.utsouthwestern.edu <histonet-request at lists.utsouthwestern.edu>
Sent: Monday, February 5, 2024 12:00 PM
To: histonet at lists.utsouthwestern.edu <histonet at lists.utsouthwestern.edu>
Subject: [EXTERNAL] Histonet Digest, Vol 243, Issue 4


CAUTION: This email originated from outside of the Marshfield Clinic Health System. Do not click links or open attachments unless you recognize the sender and know the content is safe.


Send Histonet mailing list submissions to
        histonet at lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
        https://urldefense.com/v3/__http://lists.utsouthwestern.edu/mailman/listinfo/histonet__;!!J7HzeEKFbK9hUUY!PGATEy43_xJX9a3K3Na-c7VKlFfxZ5567gwny4UONDtiVRyrbBI5aHBtxx6V0DTFWbIgFABz11BpPHqIM6HdmZ9BO2ad6vok37eGCo6_gU_9gOLs_lTLCQ$
or, via email, send a message with subject or body 'help' to
        histonet-request at lists.utsouthwestern.edu

You can reach the person managing the list at
        histonet-owner at lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific
than "Re: Contents of Histonet digest..."


Today's Topics:

   1. Re: Processing artifact - delayed start (Bacon, Charles)


----------------------------------------------------------------------

Message: 1
Date: Mon, 5 Feb 2024 13:07:41 +0000
From: "Bacon, Charles" <Charles.Bacon at baystatehealth.org>
To: Verizon wireless <someperson43 at yahoo.com>,
        "histonet at lists.utsouthwestern.edu"
        <histonet at lists.utsouthwestern.edu>
Subject: Re: [Histonet] Processing artifact - delayed start
Message-ID:
        <DM6PR02MB4265A07FD5E22F4D78A7C64A92472 at DM6PR02MB4265.namprd02.prod.outlook.com>

Content-Type: text/plain; charset="utf-8"

We have had 2 reasons why we saw processing issues like this:

1. I recently found out on our VIP 5 they did not turn the level sensors on during install. These sensors are known to error so often, Sakura tells technicians to set the default to off. All the processor can sense is pressure and time. So it may be that the formalin is not filling all the way. You are noticing this on delayed runs, but on those runs are you often stacking trays? If so, isolate the cases in the upper trays and review.

2. We found a clogged line. This happens with the formalin lines if you don?t do the hot water flush often enough. This can be an even bigger issue if you recycle and re-buffer your formalin (we do not).

Good luck!

Chuck Bacon, HTL(ASCP)CM
Supervisor Histology
Baystate Medical Center


-----Original Message-----
From: Verizon wireless <someperson43 at yahoo.com>
Sent: Friday, February 2, 2024 10:37 PM
To: histonet at lists.utsouthwestern.edu
Subject: [Histonet] Processing artifact - delayed start

Dear Histonet Members,

We have terrible processing artifact if tissue sits in the formalin-filled retort (at ambient temperature) for too long (more than 10-12 hours) before a delayed process starts. The longer the wait, the worse it looks. We have Tissue Tek VIP 5 processors, and we process luminal gastrointestinal biopsies exclusively. I've attached some photomicrographs of problem cases on the Histonet Images website (with the same topic title).
This artifact typically affects a few specimens per day (~2% or less), even though everything is done on the same processor; it may affect all tissue portions in a cassette or only some of the tissue in that cassette. Some tissue portions may only have the artifact on one half with the other half looking perfectly fine. The techs sometimes note the tissue feeling "crunchy" at the time of embedding and / or at the time of microtomy. These tissues tend to suffer greater chatter artifact and have trouble sticking to the slides. The sections look just as bad on recuts as the originals. Re-processing does not seem to help at all.
If the cassettes sit in formalin in a container outside of the processor for days before the processor is loaded (with subsequent immediate start), things look perfectly fine. When we have staff around to start the processor immediately upon loading cassettes and empty immediately upon completion, the tissue looks perfectly fine.

Our current processing program is as follows:

1. 10% Formalin, 30 minutes, ambient temp, p/v on 2. 10% Formalin, 30 minutes, ambient temp, p/v on 3. 65% Alcohol, 10 minutes, ambient temp, p/v on4. 80% Alcohol, 10 minutes, ambient temp, p/v on5. 95% Alcohol, 7 minutes, ambient temp, p/v on6. 95% Alcohol, 7 minutes, ambient temp, p/v on 7. 100% Alcohol, 10 minutes, ambient temp, p/v on 8. 100% Alcohol, 10 minutes, ambient temp, p/v on 9. Xylene, 15 minutes, ambient temp, p/v on 10. Xylene, 15 minutes, ambient temp, p/v on 11. Paraffin, 10 minutes, 60 degrees C, p/v on12. Paraffin, 10 minutes, 60 degrees C, p/v on13. Paraffin, 15 minutes, 60 degrees C, p/v on14. Paraffin, 15 minutes, 60 degrees C, p/v on

Obviously there are times we absolutely need to use delayed start. I would greatly appreciate guidance, and I'll be happy to provide any other details that might be useful.
Sincerely,
Brian Quigley MDLaboratory Director of a GI Pathology Laboratory



------------------------------

Subject: Digest Footer

_______________________________________________
Histonet mailing list
Histonet at lists.utsouthwestern.edu
https://urldefense.com/v3/__http://lists.utsouthwestern.edu/mailman/listinfo/histonet__;!!J7HzeEKFbK9hUUY!PGATEy43_xJX9a3K3Na-c7VKlFfxZ5567gwny4UONDtiVRyrbBI5aHBtxx6V0DTFWbIgFABz11BpPHqIM6HdmZ9BO2ad6vok37eGCo6_gU_9gOLs_lTLCQ$

------------------------------

End of Histonet Digest, Vol 243, Issue 4
****************************************

______________________________________________________________________
The contents of this message may contain private, protected and/or privileged information.  If you received this message in error, you should destroy the e-mail message and any attachments or copies, and you are prohibited from retaining, distributing, disclosing or using any information contained within.  Please contact the sender and advise of the erroneous delivery by return e-mail or telephone.  Thank you for your cooperation.


More information about the Histonet mailing list