Charles,
The issue is most likely the handle for the rack, The rack has a moveable handle. Make sure the handle is securely in the upright position. It has pins and holes that should line up and click into place. If it is even slightly at an angle or not securely snapped into an upright position it may bump into the side of the stain pots (and really cause a mess).
Mike