[Histonet] RE: Nuclear "Artifact"
Jennifer MacDonald
JMacDonald <@t> mtsac.edu
Wed Apr 22 17:33:35 CDT 2015
Our pathologists also complained about some of the GI biopsies looking
burnt. We tracked all of the problem cases back to one histotech. The
histotech was causing the "burn" artifact with excessive use of freezing
spray.
From: "Arbaugh, Roberta" <rarbaugh <@t> csdermatology.com>
To: "'Morken, Timothy'" <Timothy.Morken <@t> ucsf.edu>, Sue
<suetp918 <@t> comcast.net>, Lisa Roy <Royl1 <@t> LabCorp.com>
Cc: "histonet <@t> lists.utsouthwestern.edu"
<histonet <@t> lists.utsouthwestern.edu>
Date: 04/22/2015 11:32 AM
Subject: RE: [Histonet] RE: Nuclear "Artifact"
Sent by: histonet-bounces <@t> lists.utsouthwestern.edu
We have had the same problem. We come to the conclusion that it was water
droplets. We had the problem when our humidity was high in the lab, or our
weekend run. The changes we made where :
1. We no longer process over the weekend . We cannot count on our heating
and cooling in our building.
2. We place a towel and a thick piece of cardboard on top of the processor
lid.
3.We do not use the really fine mesh cassettes. We will use formalin
soaked sponges or perm papers.
4. We do not over pack cassette basket.
We had our processor check every time we saw the artifact and they could
never find a problem with the processor.
Hope this help, Roberta
-----Original Message-----
From: Morken, Timothy [mailto:Timothy.Morken <@t> ucsf.edu]
Sent: Wednesday, April 22, 2015 11:16 AM
To: Sue; Lisa Roy
Cc: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] RE: Nuclear "Artifact"
"One pathologist said it looks like the tissue has been cooked." Which
could also be drying artifact after bx, before formalin.
"The only issue is we can have two biopsies right next to one another in
the basket one looks good and one looks bad. My director also thinks it is
the processors."
Same processor but two results? Solving intermittent problems takes time
to check variables - time almost no one wants to spend to check out every
possibility. But if one variable is the same for both, and the result is
different, then most likely it is a different variable causing the
problem.
I think the other idea suggested of checking the handling of the tissue at
the source of the biopsy is more likely to shed some light on this issue.
Tim Morken
Pathology Site Manager, Parnassus
Supervisor, Electron Microscopy/Neuromuscular Special Studies Department
of Pathology UC San Francisco Medical Center
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [
mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Sue
Sent: Tuesday, April 21, 2015 4:55 PM
To: Lisa Roy
Cc: histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] Nuclear "Artifact"
OMG we are experiencing the same issue. At first it was just GI and now we
are seeing it on prostate. One pathologist said it looks like the tissue
has been cooked. The only issue is we can have two biopsies right next to
one another in the basket one looks good and one looks bad. My director
also thinks it is the processors. I had Thermo out and they could find
nothing. We changed out all the reagents and the biopsies were fine than
two days later we had some bad ones. I know in July Fisher had a formalin
recall associated to the mixture of buffer, water and formalin. We thought
that might be it but it is now almost a year later and all the bad
formalin should be gone. The histotechs say the tissue is crunchy and they
are right. I am running a test tonight of a small needle biopsy that I
made from a colon. I placed it is straight formaldehyde overnight and am
processing it on our biopsy cycle tonight. My director also wanted us to
only put three levels on our Thermo, but he wanted the middle level to
have empty baskets. I stopped that today because I think the other issue
is that the poor biopsies may be on the top level and as the reagents are
used the level changes, and also due to displacement with the middle level
being empty the reagent levels may not reach the top. We just do not have
the manpower to inspect every reagent every day, we have 6 processor and
it would take a tech all day. We actually take a digital picture when they
come out of the processor. I want to check my problems cases tomorrow. We
do not use sponges but the only other like was the PA who was wrapping the
blue paper very tight around the tissue. I really do not think this is the
issue though.. Any other insight would be greatly appreciated.
Susan T. Paturzo
TJUH
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
DISCLAIMER: The information in this message is confidential and may be
legally privileged. It is intended solely for the addressee. Access to
this message by anyone else is unauthorized. If you are not the intended
recipient, any disclosure, copying, or distribution of the message, or any
action or omission taken by you in reliance on it, is prohibited and may
be unlawful. Please immediately contact the sender if you have received
this message in error. Thank you.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list