[Histonet] Re: Ventana Ultraview users
Beth Cox
bethcoxx <@t> gmail.com
Thu Jul 10 12:56:42 CDT 2014
Beth,
We are doing exactly that same thing right now. This is changing a
complete detection system, so we need a full validation (10 pos + 10
negs for all non-prognostic markers). To avoid running 20 slides on
every antibody on our menu ($$$$), we have chosen to use a 3-prong method:
1. we bought appropriate TMAs from reputable vendors (such as
Pantomics, Inc) which will provide more than 10 Pos +10 neg for most cases.
2. we made a universal "negative" control with 12 different tissues
from our collection (provides tissues processed by us)
3. we run 1(or 2) of our control slides (bonus: also validates the
control at the same time)
This way we only run three slides for each validation to get the
required number of tissues - big difference in workload and cost!
Beth Cox, HTL/SCT(ASCP)QIHC
Pathology Solutions, Inc
__________________________________________________________________________
Message: 1
Date: Wed, 9 Jul 2014 12:16:55 -0500
From: Beth Brinegar<bbrinegarhtl <@t> gmail.com>
Subject: [Histonet] Ventana ultraview users
To:"histonet <@t> lists.utsouthwestern.edu"
<histonet <@t> lists.utsouthwestern.edu>
Message-ID:
<CAFYmSULScGQdaSdrU0dBz=55YAq-QLwsM_Z944a1hDEniW7VFA <@t> mail.gmail.com>
Content-Type: text/plain; charset=UTF-8
Hello all,
Our lab is getting ready to switch from using the iview DAB to the
ultraview - any tips for validation? How many cases did you run for each
antibody?
Thanks!
Beth Brinegar HTL(ASCP)
Anatomic Pathology Supervisor
Mercy Medical Center
Cedar Rapids, IA 52403
More information about the Histonet
mailing list