In most labs, the processor runs all night long. Someone comes in very early 
in the morning, empties the processor, starts the purge cycle, and then 
starts embedding a lot of blocks.
The tissue processor then sits, doing nothing, from after the purge in the 
early morning, until sometime in the late afternoon, when all the tissues 
are loaded in for the overnight run. That means the tissue processor is 
doing nothing for up to 12 hours during the daytime.
How about, besides the overnight run, we can set up 1 or 2 other shorter 
runs during the day, with the small biopsies.
How about - process all the large tissues overnight, but keep the little 
biopsies that you grossed all afternoon in formalin until the morning. Empty 
out the large tissues, purge the process, embed the large tissues and start 
microtoming them.
After the purge is done, put the small biopsies from the afternoon on the 
tissue processor, and process them for 1.5-2 hours (and if your processor is 
able to process half a load, do that to save on reagents). Then embed them 
and start microtoming them. Purge the processor again.
In the mean time, all morning, collect the small biopsies again. After 
lunch, short process all the small biopsies from the morning. Embed them in 
the afternoon, purge the processor again, and load up the overnight load.
If you don't have time to microtome the morning small biopsies (that you 
embedded in the afternoon), someone can microtome them in the morning the 
next day. Either with the large overnight load, or have someone else come in 
early, and while the other people are embedding the large tissue overnight 
load, they can be microtoming the small biopsies that were embedded in 
previous afternoon.
Yes, all of this will mean changes:
- staggered hours that people will be coming in
- processing, embedding and microtoming continuously throughout the day
- someone might have to microtome more than someone else, or might have to 
embed more than someone else. But if you rotate jobs around, over the 
months, everyone ends up doing the same amount of work overall.
This is a type of continuous work flow, and does lead to faster turn around 
time and efficiency.
When our lab changed to this system (we are an 1100 bed hospital, with lots 
of tissues from our ORs, from outside hospitals and clinics and doctors 
offices, so lots and lots of blocks), it took getting everyone involved - 
people accessioning, grossing, the histotechs, and the pathologists (they 
were not going to get their slides in numerical order). We have short 
cycles, the overnight long cycles, some rush cycles, and long cycles for 
breast and autopsy brain. We actually have more than 3 runs, but then we are 
working almost 24/7. During the time we were switching to continuous work 
flow, we had a few histotechs off on maternity and/or medical leaves. And we 
got a couple more clients, so the work load increased. But because of the 
continuous work flow, we were able to handle the additional work without 
having to hire anyone.
Whereas before, we would process ALL the blocks overnight, and then would 
have lots of people embedding lots of blocks first thing in the very early 
morning, and then having to put in the in order, and no one could start 
microtoming until all the blocks were embedding and in order (so some days 
the microtoming techs were sitting around with nothing to do for a time). 
Then, everyone had piles of blocks to cut, for blocks 100-150 were going to 
be cut hours and hours after everyone started microtoming blocks 1-50. Then, 
there were racks and racks of slide piling up to be stained with H&E (and at 
that point, we were still labeling after staining). So there were all these 
spots where tissue was being held up, unnecessarily:
- waiting to be processed
- waiting to be embedding
- waiting to be microtomed
- waiting to be stained
- waiting to be labeled
- then the pathologists had stacks of slides, waiting to be diagnosed
- waiting for the reports to be typed
With continuous flow, there is always some work coming through, but not 
large piles causing long waits.
It just takes rethinking how you can use the processor more efficiently, 
which will make the rest of the work more efficient. And increase 
productivity and make turn around time faster.
Peggy A. Wenk, HTL(ASCP)
-----Original Message----- 
From: Podawiltz, Thomas
Sent: Monday, January 06, 2014 12:04 PM
To: Lee & Peggy Wenk ; Deanna Leslie ; histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Soaking artifact
I agree Peggy. Just one question. What is a small histology lab to do when 
they only have one processor and cannot run separate cycles and do not have 
staffing to run short cycles throughout the day?
Tom Podawiltz HT (ASCP)
Histology Section Head/Laboratory Safety Officer.
LRGHealthcare
Laconia, NH 03246
603-524-3211 ext: 3220
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu 
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Lee & Peggy 
Wenk
Sent: Monday, January 06, 2014 8:07 AM
To: Deanna Leslie; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] Soaking artifact
The only "soaking" artifacts that I can think of would be caused by:
- soaking too long in water (minutes instead of a few seconds)
- soaking under-processed tissue in water
In both cases, the tissue is supposed to be "protected" by the wax, and if 
it is not (under-processed), or if the faced block is in water too long, the 
tissue can start re-absorbing water. The tissue then turns white and swells 
out of the block. So all that "swelled out" tissue is cut away and lost when 
the tissue is put back on the microtomy for sectioning the ribbon.
If you soak for just a few seconds, such as a gauze with water being held 
against the block on the microtome, after it has been faced, then you will 
get a little bit of water absorbed into just a few layers of cells. Just 
enough to cut 2-4 sections. And you won't see that swelling artifact.
For those of you saying - "but I have to face all the blocks, put them back 
on ice and/or water while I cut a bunch more blocks, and then go back and 
cut each block" - that is an artifact also. You have over-dehydrated your 
tissue during processing, and you are putting back the water that you should 
not have taken out. Processing is supposed to remove the unbound water (not 
attached to proteins), and some of the bound water (attached to proteins), 
and leave some of the bound water (attached to proteins) in the tissue. If 
you HAVE to soak EVERY block for more than a couple of seconds, then you are 
wasting time rehydrating and wasting time while microtoming. Cut down the 
time in the alcohols on the tissue processor, to leave a little bound water 
in the tissues. And you can NOT processing little biopsies on the same long 
processing cycle as the larger pieces of tissue (uterus, breast, etc.).
Those little biopsies will be over-dehydrated. They HAVE to be run on a 
separate cycle of much shorter time intervals (10-20 minutes in each 
solution (once fixed), instead of 45-60 minutes in each solution).
You should be able (on nearly every tissue block) to rough trim the tissue, 
and immediately start cutting ribbons. Possibly, you will need to put an ice 
cube on some of the harder tissue (cervix, uterus, bone, lens, etc.) just to 
get the paraffin hardness to match the hardness of the tissue. That being 
said, some tissues are naturally brittle or crumbly, and always need some 
water put back in the tissue, such as spleen or bloody tissue, but again, 
some wet gauze on the faced block for a few seconds should be enough time to 
get 2-4 sections. And that's all the tissue we usually need from those 
blocks. If you need more for IHC, put the wet gauze back on the faced block, 
and cut a few more sections.
Peggy A. Wenk, HTL(ASCP)SLS
-----Original Message-----
From: Deanna Leslie
Sent: Sunday, January 05, 2014 4:42 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Soaking artifact
Has anybody in histoland ever heard of this?  I have been cutting tissue for 
25 yrs and until recently I had never heard of this!
I am under contract to a facility and the supervisor there does not want 
anybody to soak their tissue or use ice!  Your are supposed to use the cold 
plate, because as I have stated soaking them causing an artifact. I have not 
disputed this because it is not my place or in my job discription as a 
traveler.  I am not even sure what it is supposed to look like or what type 
of problems it causes.
Thanks for listening!
Deanna Leslie HT ASCP
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
THIS MESSAGE IS CONFIDENTIAL.
This e-mail message and any attachments are proprietary and confidential 
information intended only for the use of the recipient(s) named above. If 
you are not the intended recipient, you may not print,distribute, or copy 
this message or any attachments.  If you have received this communication in 
error, please notify the sender by return e-mail and delete this message and 
any attachments from your computer. Any views or opinions expressed are 
solely those of the author and do not necessarily represent those of 
LRGHealthcare. 
    
    
More information about the Histonet
mailing list