[Histonet] -80 freezer question
James Watson
JWatson <@t> gnf.org
Wed Nov 20 10:18:26 CST 2013
Paula,
I do not remember exactly what study or publication I got this from, but I had to do some research a few years ago because our lab ops department did not want to have our freezer set at -80. I came up with a few reasons:
. Some proteins have been shown to break down at -60 but be stable at -80
. DNA and mRNA recovery is better at -80 vs. -60
. Some antibodies are better preserved at -80
o Our Collagen X antibody from Abcam (ab58632) requires -80 storage
Diagn Mol Pathol. 1992 Mar;1(1):73-9.
Role of the frozen tissue bank in molecular pathology.
Naber SP, Smith LL Jr, Wolfe HJ.
Source
Department of Pathology, Tufts University School of Medicine, New England Medical Center Hospitals, Boston, MA 02111.
Abstract
The new discipline of molecular pathology requires that high-quality, intact genomic DNA, mRNA, and proteins be available from frozen tissue samples. It is now necessary for pathology laboratories to establish consistent guidelines for the preparation and storage of frozen tissue samples in order to have properly preserved tissues available for diagnostic molecular techniques. Maintaining a frozen tissue bank requires a pathologist to oversee this program and to integrate it into the routine surgical pathology activities. A member of the laboratory technical staff can serve as a tissue bank coordinator and have responsibility for preparation of tissue samples, their systematic storage and retrieval, and routine maintenance of equipment and supplies. Tissue sampling must be done as soon as possible after excision of the specimen and is the responsibility of a qualified pathologist. The samples may be snap frozen without cryoprotection at -78 degrees C or colder for subsequent use in procedures requiring the extraction of genomic DNA, mRNA, or protein. To preserve tissue architecture and cytologic features for immunohistochemistry and in situ hybridization, the tissue should be frozen at -78 degrees C or colder with a cryoprotectant such as OCT. Long-term storage of the frozen tissue is recommended at -140 degrees C or colder in a locked liquid nitrogen freezer, and the record of sample inventory can easily be kept in a computerized database. Tissues sampled and stored under these conditions have been used successfully in a wide variety of molecular techniques. In addition to malignant tumor tissue, samples from benign lesions and normal tissues should be frozen.(ABSTRACT TRUNCATED AT 250 WORDS)
PMID:
1342956
[PubMed - indexed for MEDLINE]
James Watson HT ASCP
GNF Genomics Institute of the Novartis Research Foundation
Tel 858-332-4647
Fax 858-812-1915
jwatson <@t> gnf.org
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Paula Sicurello
Sent: Tuesday, November 19, 2013 2:23 PM
To: HistoNet
Subject: [Histonet] -80 freezer question
Dear Netters,
What is the rationale for storing frozen patient samples at -80? Is there a range that keeps the sample viable but not as cold?
Our -80 (which we just inherited) is struggling and I would like to raise the temperature a smidge to ease the stress on the compressor.
Thanks!
Paula
--
Paula Sicurello, HTL (ASCP)
Supervisor, Clinical Electron Microscopy Laboratory Duke University Health System Rm.#251M, Duke South, Green Zone Durham, North Carolina 27710
P: 919.684.2091
HIPAA Privacy Notification: This message and any accompanying documents are covered by the Electronic Communications Privacy Act, 18 U.S.C. 2510-2521, and contain information intended for the specific individual (s) only. This information is confidential. If you are not the intended recipient or an agent responsible for delivering it to the intended recipient, you are hereby notified that you have received this document in error and that any review, dissemination, copying or the taking of any action based on the contents of this information is strictly prohibited . If you have received this communication in error, please notify us immediately by e-mail, and delete the original message.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet