[Histonet] RE: tol blue and mast cells
Elizabeth Chlipala
liz <@t> premierlab.com
Fri Jul 12 10:21:50 CDT 2013
Richard
We use 0.04% T. Blue in an acetate buffer that is pH'd to 4.0. Stain for about 5 minutes, rinse in water, let air dry and then coverslip in xylene. You will loose the metachromatic staining if you run the slides through alcohol. I have an SOP if you would like to see one.
Liz
Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 881-0763 cell
(303) 682-9060 fax
liz <@t> premierlab.com
Ship to address:
Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504
________________________________________
From: histonet-bounces <@t> lists.utsouthwestern.edu [histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Edwards, Richard E. [ree3 <@t> leicester.ac.uk]
Sent: Friday, July 12, 2013 9:17 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] tol blue and mast cells
Hi All, your thoughts on the best toluidine blue method for mast cells?, many thanks.
Richard Edwards
University of Leicester
U.K.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list