Hey Deloris,

I would have the dispensing nozzles on the XT checked, could be it is not 
getting rinsed properly.  Had that issue with my Ultra not to long ago.  No 
warnings or errors showed up, just a clogged nozzle tip.

Tim

-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu 
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Deloris 
Carter
Sent: Monday, November 05, 2012 5:16 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Background staining on H pylori

Hi,
I'm getting a lot of background staining on my HP's.  We use a Ventana 
BenchmarkXT with ultraView DAB.  The problem seems to be escalating of late, 
and I'm not sure why.  We use Hollande's on our GI biopsies, and run them on 
a shorter run.  Nothing has changed in the processing of the specimens, or 
the IHC procedure.  The antibody dispenser is a newer lot, but not brand 
new, as we get the larger size due to the high volume of HP's we run. Any 
ideas?