CAP has a "suggestion" for a validation protocol BUT the most important thing is the acceptance of your processed tissues by the pathologist. S/he will be the ultimate responsible in the validation process.
Ideally, you should run parallel samples with the instrument you are using now and the Peloris. The number varies. You will also have to do IHC in some of the samples run with the Peloris. Select those IHC of greater importance and most frequent use. Ask your pathologist which s/he wants.
In all these validations do not try to do it by yourself, always ask the pathologists what s/he wants to do.
At first glance validation seems to be a daunting task but it is not if you follow your pathologists' indications.
Document the whole process and make sure to have the pathologist's signature on the evaluation of the slides and procedures. Archive those blocks and slides for future reference.
That is how I did all my validations even when I was developing new processing protocols.
René J.
--- On Thu, 3/29/12, Michelle Wahlsten <mwahlsten <@t> yahoo.com> wrote:
From: Michelle Wahlsten <mwahlsten <@t> yahoo.com>
Subject: [Histonet] Leica Peloris II processor validation
To: "Histonet <@t> lists.utsouthwestern.edu" <Histonet <@t> lists.utsouthwestern.edu>
Date: Thursday, March 29, 2012, 9:30 PM
Hi everyone,
I'm currently in process of validating the Peloris II for our lab, my first instrument validation so I'm a bit lost as to what specifically I need to do and document. Other than running specimens through the specific protocols to verify the quality of the specimen with a pathologist's review and/or sign-off, I feel that there is more that I need to do but I'm not sure what that is. Can anyone share with me their experience of the validation process and/or suggestions as to what else I need or what I'm missing??
In addition to the instrument validation itself, we also need to validate IHC use on the Peloris, per one of our pathologist's. How many antibodies, controls or panels should we be doing to satisfy the 'validation' process for IHC?? Any information will be helpful, thank you so much!!