[Histonet] Background in FFPE glandular tissue sections
JMyers1 <@t> aol.com
JMyers1 <@t> aol.com
Fri Sep 30 10:02:49 CDT 2011
Mesruh:
The easiest method to address such is a problem is to apply a serum-free
(e.g. casein-based) protein blocking agent, which can be obtained from a
variety of vendors. This technique is particularly effective when the reagent
is allowed to incubate with the specimen for 5 to 10 minutes at RT, and
then, rather than rinsing it off the slide with buffer, it is drained off or
blown off; that way, when the primary antibody solution is applied
(immediately after this removal step), the primary has to 'work through' the
residual blocking agent, resulting in the binding of only high-avidity/affinity
antibodies to the target antigen.
Good Luck,
Joe Myers, M.S., CT(ASCP)QIHC
------------------------------
Message: 3
Date: Thu, 29 Sep 2011 13:10:58 -0400
From: mesruh turkekul <turkekul <@t> gmail.com>
Subject: [Histonet] Background in FFPE glandular tissue sections
during IHC
To: _histonet <@t> lists.utsouthwestern.edu_
(mailto:histonet <@t> lists.utsouthwestern.edu)
Dear Histonetters,
When I do IHC on FFPE tissue sections. Most of time there is background due
to trapping of IgG/detection reagents probably in the connective tissue
fibers, mucins or other secretions of the glandular tissue. Especially
prostate and mammary gland. Any tips to get rid of that kind of background?
Do you know any treatment to block mucins or connective tissue fibers from
nonspecifically and electrostatically attracting IgGs and detection
reagents?
Regards,
Mesruh Turkekul
mskcc.org