Reply to cryostat issue:

If the cryostat has a foot pedal attachment, like the Leica CM3000, sometimes the chamber will not cool if the foot pedal is not plugged in.  

Best,
Joyce Kwan
Elan Pharmaceuticals


From: mohamed abd el razik <k84as <@t> yahoo.com>
Subject: [Histonet] cryostat help please!
To: histonet <@t> lists.utsouthwestern.edu
Date: Saturday, March 20, 2010, 9:30 AM


hi all histoneters 
i have learned alot from your posts and i need any suggestion about a cryostat that was
working perfectly until the power cable is pluged off accidently. then after about 2 days i put it again in the power and all is working with its motor sound but no refregration and room temp. not decrease under 20 degrees!!! i let it for aweek but no change!  unfortene the technical service may take long time till it come to see the proplem as it is in a national university ( routine work) and i hope i could do that. any suggestion please?


********************************************************