[Histonet] New CAP question ANP.22760

Mike Pence mpence <@t> grhs.net
Fri Jun 18 13:28:38 CDT 2010


As I have found over the years the opinions and interpretation of CAP
intent is differing. Everyone has how they see the question and think it
to best be answered. I am not sure why it is not acceptable for the lab
to run a control on the new lot and make sure it works and that is the
QC/QA. The pathologist is the one that determines if the stain is
acceptable for diagnosis. Some of the suggestions would be very time
consuming and costly to do each new lot#.

-----Original Message-----
From: Vickroy, Jim [mailto:Vickroy.Jim <@t> mhsil.com] 
Sent: Friday, June 18, 2010 1:06 PM
To: Mark Tarango; Mike Pence
Cc: histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] New CAP question ANP.22760


I'm kind of getting in on the middle of this but is anybody else doing
this with prep kit stickers?  I have not seen the new question but has
anyone talked to CAP to get a clarification of what they mean on this
question?

James Vickroy BS, HT(ASCP)

Surgical  and Autopsy Pathology Technical Supervisor
Memorial Medical Center
217-788-4046


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Mark
Tarango
Sent: Friday, June 18, 2010 12:27 PM
To: Mike Pence
Cc: histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] New CAP question ANP.22760

You'll have to use prep kit stickers and duplicate protocols to do it,
but it is possible.  You just need to copy the protocol and save it as
another number, then change the primary antibody to a prep kit sticker
save again and then put that sticker on the dispenser.  Then you need to
print stickers for both protocols and stick them on two controls slides
and run.

I admit its a little bit of a pain.

Mark

On Fri, Jun 18, 2010 at 9:41 AM, Mike Pence <mpence <@t> grhs.net> wrote:

>
> I don't think I can do this with the automated system we are currently

> using. Ventana. Does any other Ventana users know if you can do this 
> in "parallel"
>
> Mike
> -----Original Message-----
> From: histonet-bounces <@t> lists.utsouthwestern.edu
> [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Ellen 
> Yee
> Sent: Thursday, June 17, 2010 7:21 PM
> To: Laurie Colbert
>  Cc: histonet <@t> lists.utsouthwestern.edu
> Subject: RE: [Histonet] New CAP question ANP.22760
>
>
> Sorry, I should have included it.
>
> ANP.22760  Are new lots of antibody and detection system reagents 
> tested in parallel with old lots?  (NOTE: New lots of primary antibody

> and detection system reagents must be compared to the previous lot 
> using an appropriate panel of control tissues.)
>
> Ellen Yee
>      _____
>
>  From: Laurie Colbert [mailto:laurie.colbert <@t> huntingtonhospital.com]
> To: Ellen Yee [mailto:eyee <@t> dpmginc.com]
> Sent: Thu, 17 Jun 2010 08:47:38 -0700
> Subject: RE: [Histonet] New CAP question ANP.22760
>
> Can you give us the wording of that question/checklist item? Laurie
>
> -----Original Message-----
> From: histonet-bounces <@t> lists.utsouthwestern.edu
> [mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Ellen 
> Yee
> Sent: Wednesday, June 16, 2010 10:10 PM
> To: histonet <@t> lists.utsouthwestern.edu
> Subject: [Histonet] New CAP question ANP.22760
>
> How are IHC labs complying with this question? What is considered an 
> appropriate panel of control tissues? What do you stain to test your 
> detection systems?
>
> Ellen Yee
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu 
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
>
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu 
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
>
>
> _______________________________________________
> Histonet mailing list
> Histonet <@t> lists.utsouthwestern.edu 
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
>
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

This message (including any attachments) contains confidential
information intended for a specific individual and purpose, and is
protected by law. If you are not the intended recipient, you should
delete this message. Any disclosure, copying, or distribution of this
message, or the taking of any action based on it, is strictly
prohibited.





More information about the Histonet mailing list