Brandi-

The flush should be with warm water (you do not need hot), and should be for the fixative station(s) and the first alcohol.  You only need 1 minute in each, just enough to get the solution to pump in.

Joe Saby, BA HT




________________________________
From: Anthony Reilly <Tony_Reilly <@t> health.qld.gov.au>
To: Brandi Higgins <brandihiggins <@t> gmail.com>; histonet <@t> lists.utsouthwestern.edu
Sent: Mon, April 5, 2010 10:36:42 PM
Subject: Re: [Histonet] Cleaning VIP Processor containers

Hi Brandi

Do you do hot water flushes? If not the film will be a build up of the buffer salts from your formalin.  To flush replace your formalin containers with hot water and write a program to run each container for 10-15 minutes.  This will not only clean the retorts but the fluid lines as well.  This should be done weekly or at least monthly if time is an issue.

regards
Tony





Tony Reilly  B.Sc. , M.Sc.
Chief Scientist, Anatomical Pathology
Pathology Queensland-PA Laboratory
_________________________________________________
Clinical and Statewide Services Division| QueenslandHealth

Level 1, Building 15,Princess Alexandra Hospital
Ipswich Road,WOOLLOONGABBA  Qld4102
Ph: 07 3176 2412
Mob: 0402 139411
Fax: 07 3176 2930
Email: tony_reilly <@t> health.qld.gov.au
Web:  www.health.qld.gov.au/qhcss/



********************************************************************************
This email, including any attachments sent with it, is confidential and for the sole use of the intended recipient(s). This confidentiality is not waived or lost, if you receive it and you are not the intended recipient(s), or if it is transmitted/received in error.
Any unauthorised use, alteration, disclosure, distribution or review of this email is strictly prohibited.  The information contained in this email, including any attachment sent with it, may be subject to a statutory duty of confidentiality if it relates to health service matters.
If you are not the intended recipient(s), or if you have received this email in error, you are asked to immediately notify the sender by telephone collect on Australia +61 1800 198 175 or by return email.  You should also delete this email, and any copies, from your computer system network and destroy any hard copies produced.
If not an intended recipient of this email, you must not copy, distribute or take any action(s) that relies on it; any form of disclosure, modification, distribution and/or publication of this email is also prohibited.
Although Queensland Health takes all reasonable steps to ensure this email does not contain malicious software, Queensland Health does not accept responsibility for the consequences if any person's computer inadvertently suffers any disruption to services, loss of information, harm or is infected with a virus, other malicious computer programme or code that may occur as a consequence of receiving this email.
Unless stated otherwise, this email represents only the views of the sender and not the views of the Queensland Government.
**********************************************************************************

_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



      


More information about the Histonet mailing list