I've worked with 4 of these units, two had the problem, two didn't. Long story short, the problem won't go away until you have the Peltier unit replaced. That is not normal ice buildup.
Laurie
------Original Message------
From: "Demarinis, Carolyn" <cdemarinis <@t> SARATOGACARE.ORG>
Date: Tue Jun 23, 2009 -- 10:29:08 AM
To: HISTONET <@t> PATHOLOGY.SWMED.EDU
Subject: [Histonet] Leica CM1850uv
We purchased a Leica CM1850uv cryostat in 2007 and we constantly have
problems with buildup of ice
on the specimen metal bar. This causes the tissue to stick to the
stationary heat extractor which causes
significant problems cutting frozen sections. The engineer has serviced
the machine several times, but
recognizes this as "normal ice buildup".
We manually defrost the unit 2-3 times a day, but it does not seem to
help.
Is anyone else experiencing the same problem, and if so, what are they
doing about it? Thank you.
This e-mail communication and any attachments may contain
confidential and privileged information for the use of the
designated recipients named above.
If you are not the intended recipient, you are hereby notified
that you have received this communication in error and that
any review, disclosure, dissemination, distribution or copying
of it or its contents is prohibited. If you have received this
communication in error, please notify Saratoga Hospital
immediately by e-mail at privacy <@t> saratogacare.org and
destroy all copies of this communication and any attachments.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list