[Histonet] Auto-IHC Staining Issues
Bull, Jennifer L.
Jennifer.Bull <@t> northwestpathology.com
Thu Jan 22 11:18:37 CST 2009
Paula-
It's a hard problem to diagnose with out knowing what platform you are using, but here is what I have learned from our experience-
- We had the same trouble on our Ventana Benchmark and after much hair pulling and troubleshooting, found it to be the slides we were using. The only slides that Ventana recommends for use on their stainer are the Erie superfrost. This fixed the problem for us. We have since found that the Bond360 slides from TruScientific work equally as well.
- We also found that we had similar problems on Dako's Autostainer. This time it was due to the instrument not being completely level. We also learned that when transferring the slides to buffer after the retrieval process, you need to be quick or they can dry out, causing artifact or patchy staining effect. This too was easily solved by using a tween buffer or by using different racks to transfer the slides.
I hope this is of help to you, I know how frustrating troubleshooting staining issues can be!
Jennifer Bull
jennifer.bull <@t> nwpathology.com