[Histonet] Benchmark XT

Joe Nocito jnocito <@t> satx.rr.com
Wed Apr 8 16:08:02 CDT 2009


another issue could be the mixer blowing too hard. I've had that happen too.

JTT
----- Original Message ----- 
From: "Kari Bradshaw" <kbradshaw <@t> lcpath.com>
To: "Dana Spears" <dspears <@t> mmci.org>; <histonet <@t> lists.utsouthwestern.edu>; 
"Phyllis Thaxton" <dchihc <@t> yahoo.com>
Sent: Wednesday, April 08, 2009 2:13 PM
Subject: RE: [Histonet] Benchmark XT


We just recently went through a similar problem with irregular staining,
light staining, sometimes no stain of any kind including
counterstain,tissue falling off, and a handfull of false negative
patient tissues, but positive controls (on the same slide). With two
XT's running continuously it taken several days and everyone's help
isolating the problem....not to mention several different lots numbers
of slides....yikes! Next time I'll blame the slides right off the bat.

Kari L. Bradshaw HT(ASCP)
Laboratory Manager
Lower Columbia Pathologists
(360)425-5620
kbradshaw <@t> lcpath.com


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Dana
Spears
Sent: Wednesday, April 08, 2009 9:24 AM
To: histonet <@t> lists.utsouthwestern.edu; Phyllis Thaxton
Subject: Re: [Histonet] Benchmark XT

I would definitely check your slides - we have found that when we get a
bad lot of charged slides, things start washing off on the XT -
regardless of detection, antibodies, whatever.  It was happening to us
on some tissue types and not on others, but a new batch of charged
slides did the trick.

I've never had to dry 60 min in the oven - 20 does it for us when we
aren't having slide issues.

Also, if you are using any Sta-on or any adhesive in your waterbath with
the charged slides it can cause the slides to sort of repel the tissue
and negate the "charged" effect.... again certain tissues will
fall off, others stay on....

Dana Spears, HTL(ASCP)
Laboratory Manager
Methodist Medical Center
(309) 672-4930 (office)
(309) 255-7214 (cell)
(309) 279-3768 (fax)
dspears <@t> mmci.org

>>> Phyllis Thaxton <dchihc <@t> yahoo.com> 4/8/2009 8:43:36 AM >>>

Is anyone out there having problems with tissue washing off the slides
using the Ultraview kit from Ventana?

Mainly we are having problems with  needle biopsies washing off
(prostate, liver). Fixation and processing is always the same 4-6  hours
fixation, 3 hour biopsy processing run on VIP processor, cut no thicker
than 4 microns, airdried at least 30 minutes, baked at 59-60 for at
least one hour prior to staining. The instrument's calibrations have
been checked and are fine.

We never had this problem with the iView kit using the Nexes, doing
pretreatments offline.

Any help, ideas, info will be appreciated.

Thanks,
Phyllis Thaxton HT(ASCP)QIHC
DCH Regional Medical Center
Tuscaloosa, AL 35401



_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



NOTICE: This message is a PRIVATE communication. This e-mail may contain
confidential or proprietary information that may be considered legally
privileged. It is intended only for the named recipient(s). If an
addressing or transmission error has misdirected the e-mail, please
notify the author by replying to this message. If you are not the named
recipient, you are not authorized to use, disclose, distribute, copy,
print, or rely on this e-mail, and should immediately delete it from
your computer system.

Thank you for your assistance with this matter.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet




_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet 




More information about the Histonet mailing list