[Histonet] Benchmark XT
Dana Spears
dspears <@t> mmci.org
Wed Apr 8 11:23:49 CDT 2009
I would definitely check your slides - we have found that when we get a
bad lot of charged slides, things start washing off on the XT -
regardless of detection, antibodies, whatever. It was happening to us
on some tissue types and not on others, but a new batch of charged
slides did the trick.
I've never had to dry 60 min in the oven - 20 does it for us when we
aren't having slide issues.
Also, if you are using any Sta-on or any adhesive in your waterbath
with the charged slides it can cause the slides to sort of repel the
tissue and negate the "charged" effect.... again certain tissues will
fall off, others stay on....
Dana Spears, HTL(ASCP)
Laboratory Manager
Methodist Medical Center
(309) 672-4930 (office)
(309) 255-7214 (cell)
(309) 279-3768 (fax)
dspears <@t> mmci.org
>>> Phyllis Thaxton <dchihc <@t> yahoo.com> 4/8/2009 8:43:36 AM >>>
Is anyone out there having problems with tissue washing off the slides
using the Ultraview kit from Ventana?
Mainly we are having problems with needle biopsies washing off
(prostate, liver). Fixation and processing is always the same 4-6 hours
fixation, 3 hour biopsy processing run on VIP processor, cut no thicker
than 4 microns, airdried at least 30 minutes, baked at 59-60 for at
least one hour prior to staining. The instrument's calibrations have
been checked and are fine.
We never had this problem with the iView kit using the Nexes, doing
pretreatments offline.
Any help, ideas, info will be appreciated.
Thanks,
Phyllis Thaxton HT(ASCP)QIHC
DCH Regional Medical Center
Tuscaloosa, AL 35401
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
NOTICE: This message is a PRIVATE communication. This e-mail may contain confidential or proprietary information that may be considered legally privileged. It is intended only for the named recipient(s). If an addressing or transmission error has misdirected the e-mail, please notify the author by replying to this message. If you are not the named recipient, you are not authorized to use, disclose, distribute, copy, print, or rely on this e-mail, and should immediately delete it from your computer system.
Thank you for your assistance with this matter.
More information about the Histonet
mailing list