I was taught that a true negative control should not be elimination of the
primary antibody (NULL control) but use the immunoglobulin matching the host
of the primary antibody and at the same concentration as the primary
antibody. Some people use nonimmune serum, but we prefer using IgG, IgG
isotype, or IgM. I also found out the hard way, in a research project,
that using non immune normal serum was unacceptable to reviewers, and the
whole, incredibly tedious and difficult immunogold staining had to be
repeated before a manuscript was accepted. A very hard lesson, and now
there is a collection of IgG or isotype controls for all immunostaining
projects available.
This immunoglobulin negative control can be used on a patient tissue. The
Ig negative control is there to evaluate nonspecific staining in that
particular tissue. If background occurs, then a source test may be
necessary, and with elimination of a primary antibody, that is the first
step in a background source test.
There is an excellent discussion of negative controls, tissue or reagent
controls, in the freebie, Dako's Education Guide, Immunohistochenisal
staining methods, Fourth Edition and available as pdf on Dako Website.
There are several ways of setting up a negative or positive control. Their
comment on page 114 of this manual, verbatim, "Finally, the diluent itself
may be used as an alternative, which, however is neither efficient nor
desirable" leads me to think the immunoglobulin control is more acceptable.
Gayle M. Callis
HTL/HT/MT(ASCP)
Bozeman MT 59715
-- Original Message -----
From: "Weems, Joyce" <JWeems <@t> sjha.org>
To: "Mike Pence" <mpence <@t> grhs.net>; "Sarah Kolekamp"
<kolekams <@t> trinity-health.org>; <histonet <@t> lists.utsouthwestern.edu>
Sent: Wednesday, April 30, 2008 11:12 AM
Subject: RE: [Histonet] Re: Negative Control Tissue
This has always been confusing, but, if I understand correctly, the
negative control is the same patient tissue without primary antibody.
Use a negative for each detection system used if you have multiple
antibodies on the same block.
Joyce
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Mike
Pence
Sent: Wednesday, April 30, 2008 1:05 PM
To: Sarah Kolekamp; histonet <@t> lists.utsouthwestern.edu
Subject: RE: [Histonet] Re: Negative Control Tissue
I think you would have to use the same tissue you use for your + control
to show that it "is working/staining" and "is not working/staining"
At least that is how we do it.
Mike
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Sarah
Kolekamp
Sent: Wednesday, April 30, 2008 11:06 AM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Re: Negative Control Tissue
I have a questions about negative control tissue, does anyone know what
the most universal tissue would be that we could use to act as our
negative control tissue. We are looking for the tissue that could be
used with the widest array of antibodies.
Thanks!
Sarah Kolekamp HTL (ASCP)
Lead Histotechnologist
Saint Mary's Health Care
Phone Number (616) 752-6155
Fax Number (616) 774-0628
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
Confidentiality Notice:
This email, including any attachments is the
property of Catholic Health East and is intended
for the sole use of the intended recipient(s).
It may contain information that is privileged and
confidential. Any unauthorized review, use,
disclosure, or distribution is prohibited. If you are
not the intended recipient, please reply to the
sender that you have received the message in
error, then delete this message.
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
More information about the Histonet
mailing list