[Histonet] MICROWAVES

Phil McArdle PMcArdle <@t> ebsciences.com
Thu Dec 20 14:57:17 CST 2007


Hi:

This response is from a microwave vendor, so you've been forewarned! :-)

That said, it seems analogous ISO-9000 certification: each lab is 
allowed to make its own determination as to the meaning of 
"periodically;" the procedure should be documented and assigned AND (and 
here's the key part) adhered to and logged! I've seen some labs with 
very nicely thought out and documented procedures, who got written up 
because they didn't follow their own procedures.

When CAP began their microwave leakage measurement requirement, it 
specified "at least" yearly, and EBS recommended quarterly. (Arbitrary? 
Maybe.) As for reproducibility, I feel that yearly is not enough, daily 
or weekly is probably excessive, but depending upon your lab's volume, 
monthly or quarterly should be sufficient, unless anything unusual is 
noted (blown fuse, etc.).

Hope this helps!

Phil McArdle
-- 
Phil McArdle
Microwave Product Manager

Energy Beam Sciences, Inc.
29-B Kripes Rd.
East Granby, CT 06026

Tel:  800.992.9037 x 341
Mobile: 860.597.6796
Fax: 860.653.0422

pmcardle <@t> ebsciences.com
www.ebsciences.com

I skate to where the puck is going to be, not to where it's been.
- Wayne Gretsky

You must be the change you want to see in the world.
- Mahatma Gandhi

NOTE: This message, together with any attachments, is intended only for 
the use of the individual or entity to which it is addressed and may 
contain information that is legally privileged, confidential and exempt 
from disclosure. If you are not the intended recipient, however, there's 
not a lot I can do about it, and it was probably my mistake anyway. So 
please do the right thing and make this e-mail go away. Thank you.


Janice Mitchell wrote:
 > Cap requests "microwave devices be periodically monitored for
 > reproducibility" what is considered periodically?
 >                       Thanks, Janice
 >
 > _______________________________________________
 > Histonet mailing list
 > Histonet <@t> lists.utsouthwestern.edu
 > http://lists.utsouthwestern.edu/mailman/listinfo/histonet






More information about the Histonet mailing list