[Histonet] re:help MVP/Ventana tissue processors

Luck, Greg D. LuckG <@t> empirehealth.org
Sat Jun 10 16:17:05 CDT 2006


To Leroy et.al.,

Try the code numbers 762 (and press enter on the keypad).  I have both
an older RMC instrument as well as a newer Ventana Renaissance.  I can
access the service diagnostics of both instruments with this code.  I
use this feature frequently to check that my remote alarms are
functioning.  The way I easily remember this code sequence is that 762
on your telephone keypad is (you guessed it) "RMC".  Contact me if you
need any help with your processor as both I and one of our biomed techs
have a lot of experience with these instruments.  Good luck, Greg

Greg Luck, BS, HT(ASCP)
Anatomic Pathology Supervisor
Deaconess Medical Center
800 W. 5th Ave
Spokane, WA 99204
Phone 509.473.7077
Fax 509.473.7133
luckg <@t> empirehealth.org
www.deaconessmc.org


-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of
RCHIOVETTI <@t> aol.com
Sent: Saturday, June 10, 2006 6:42 AM
To: int09018 <@t> alphahunt.com; histonet <@t> lists.utsouthwestern.edu
Subject: Re: [Histonet] re:help needed for older tissue processor

In a message dated 6/9/2006 8:10:45 PM US Mountain Standard Time,
int09018 <@t> alphahunt.com writes:

> Hi, I am looking for leads or any information on the RMC MVP 1 tissue 
> processor also know as the Instrumentation Laboratory MVP.  It is 
> Model 1505.  My biggest problem is that I need the code that allows me

> to go into the calibration and service screen on the menu.

LeRoy,

Call Ventana Medical Systems, (800) 227-2155  and ask for Tom Kennedy.
He should be able to help you with the service code.

Good Luck!

Bob 

Robert (Bob) Chiovetti, Ph.D.
Owner
The Microscope Works
Arizona's Microscopy Resource
132 North Elster Drive
Tucson, AZ 85710-3212 USA
Tel./Fax 520-546-4986
Manufacturers' Representative
Member, Arizona Small Business Association - ASBA
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet