[Histonet] uneven sections on slides

Jennifer MacDonald JMacDonald <@t> mtsac.edu
Thu Mar 17 15:06:59 CST 2005


Multi levels of focus can also be attributed to sections that are too 
thick.

Jennifer MacDonald




RCHIOVETTI <@t> aol.com 
Sent by: histonet-bounces <@t> lists.utsouthwestern.edu
03/17/2005 08:54 AM

To
Julie.Sanders <@t> med.va.gov, Histonet <@t> lists.utsouthwestern.edu
cc

Subject
Re: [Histonet] uneven sections on slides






In a message dated 3/17/2005 9:25:56 AM US Mountain Standard Time, 
Julie.Sanders <@t> med.va.gov writes:

> In the past few days one our pathologists has been complaining about 
> "uneven
> sections."  In other words, he's having to focus up and down a lot 
because
> the sections aren't flat on the slide. 


Julie,

It sounds to me like the problem may lie in the microtome.  It could be a 
type of "chatter" in the sections.  They may indeed not be flat on the 
slide, or 
they could be of uneven thickness.

This is almost always because something is loose on the microtome, such as 

the specimen clamp or something in the knife holder, and the main offender 
is 
usually the knife holder.

A quick check would be to cut the same block on different microtomes, then 

ask the doc to evaluate the sections.  See if you can pinpoint the problem 
to a 
specific instrument. 

If you find you have a problem with one microtome, be sure to check out 
the 
knife holder and the clamping mechanism for the blade.  It could be 
something 
as simple as paraffin debris in the blade clamping mechanism, or maybe the 

clamping pressure needs to be adjusted or the mechanism's springs need to 
be 
replaced.

Just my two cents' worth.  Hope it helps!

Cheers,

Bob