At my former institution we were doing a large volume of immuno's and
would every 3 to 6 months start having inconsistant staining and have to
get parts replaced on the probe assembly. That usually fixed it, so we
got the the point where every 3 months we had them in for a PM. The
components that were problematic were part of the normal PM.
One time the problem continued and multiple things were replaced (this
was on a 4 or 5 year old machine). It turned out that it was the power
supply transformer in the instrument. It would occassionally kick out
an odd voltage and thus the pipet would pick up too little reagent.
In spite of that problem I personally still prefer the Dako.
Ross M Stapf
Histopathology Manager
Baylor University Medical Center
3500 Gaston Ave.
Dallas, TX 75246
214-820-2465
214-820-4110 fax
RossS <@t> baylorhealth.edu
-----Original Message-----
From: histonet-bounces <@t> lists.utsouthwestern.edu
[mailto:histonet-bounces <@t> lists.utsouthwestern.edu] On Behalf Of Hopkins,
Karen
Sent: Monday, February 14, 2005 12:15 PM
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Dako Autostainer
Hi We have had a problem with our IHC autostainer and I was wondering if
anyone else has experienced the same problem. We had a Dako Autostainer
and over time (about three years) found that we had intermittent
problems with slides not being stained. It started with slides in the
first and second slide position (the positive control was not staining)
when we had a large run but eventually we noticed it at other positions,
spiratically. We have another instrument now, since Nov and the other
day experienced the same problem. We ran three sets of known controls
(we send one to each of the facilities we stain for) and one of the
three was positive, the other two were not. We were using a RTU ER
antibody. ???? Also, the spec sheet from Dako states we should run
the ER with Env or Lsab2. We get better results with ENV+. The stain
with ENV is extremely pale. Any comments or suggestions?
_______________________________________________
Histonet mailing list
Histonet <@t> lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
This e-mail, facsimile, or letter and any files or attachments transmitted with it contains information that is confidential and privileged. This information is intended only for the use of the individual(s) and entity(ies) to whom it is addressed. If you are the intended recipient, further disclosures are prohibited without proper authorization. If you are not the intended recipient, any disclosure, copying, printing, or use of this information is strictly prohibited and possibly a violation of federal or state law and regulations. If you have received this information in error, please notify Baylor Health Care System immediately at 1-866-402-1661 or via e-mail at privacy <@t> baylorhealth.edu. Baylor Health Care System, its subsidiaries, and affiliates hereby claim all applicable privileges related to this information.