Dear Tess,
Are you sure the cryochamber is below -30ºC or just the specimen is at that temperature?? As your problem indicates that the knife & anti-roll plate are too warm, if the specimen is at -30ºC. With the correct temperatures, mammary gland should be easy to section.
Best Regards
Alan Bright
Bright Instrument Co.Ltd.
St Margaret's Way
Huntingdon
Cambridgeshire
PE29 6EU
England
Tel No:+44 (0)1480 454528
Fax No:+44 (0)1480 456031
Email: abright <@t> brightinstruments.com
Web Site: www.brightinstruments.com
-----Original Message-----
From: Tessa Murray [mailto:tessajmurray <@t> hotmail.com]
Sent: 10 August 2005 21:50
To: histonet <@t> lists.utsouthwestern.edu
Subject: [Histonet] Cryostat sectioning of mammary gland
I'm looking for some advice on cryosections. We are trying to collect
frozen sections of mammary gland tissue and are experiencing problems
- we do a lot of paraffin work here so the cryostat technique is quite
new to us. We find that the knife cuts the OCT fine, but when it hits
tissue it does not cut so we get sections with tissue shaped holes in
them. We've tried changing the block and knife temp (we have tried
altering them between -30 to -35C) but nothing seems to help. Any
advice appreciated...
tess