[Histonet] Background with tape transfer slides, IHC
Gayle Callis
gcallis <@t> montana.edu
Tue Apr 12 14:08:01 CDT 2005
Patsy,
I have not had a problem. I do add Tween 20 to all buffers and diluents
and we only use 1/2X slides. We find these hold bone perfectly but we do
double flash to ensure no lift off. The Tween is 0.05% but you can try
and increase concentration higher, 0.2% or so, maybe even higher than
that. You may be experiencing ionic interaction of chromogen to the
polymer, commonly occurs with plus charge slides also, i.e. ring around the
section only you have tinted polymer.
I am assuming you are using DAB and that is the background??
I am doing hrp IHC on tape transfer sections and I am experiencing a problem
with unwanted staining on the coating on the polymer coated slides. Does
anyone else out there do IHC on the tape sections experience this problem?
Please suggest a solution to this.
Thank you,
Patsy